我有一个客户参与校准已经安全,我试图使用手册中概述的程序绕过安全性,但它似乎不起作用。
该手册说明了底部标有“cal unlock”的两个金属垫的外壳和短路,然后输入“任何安全代码”,瞧! 您已解锁并准备好根据需要输入新代码。 然而,我正在研究的装置说“不匹配”,就像它没有跳过校准解锁一样,它仍然可以防止校准。 任何帮助赞赏。 以上来自于谷歌翻译 以下为原文 I have a customer piece in for calibration that has been secured, and I'm trying to bypass the security using the procedure outlined in the manual, but it doesn't seem to work. The manual says to take the case of and short the two metal pads marked "cal unlock" on the bottom side, then enter "any security code" and voila! you're unlocked and ready to enter a new code as needed. The unit I'm working on, however, says "no match" just like it does without jumping the cal unlock and it remains secured against calibration. Any help appreciated. |
1个回答
这个程序记录很少,我会修复它。
您必须将显示器置于显示HP000000或AT000000的模式。 然后短接矩形垫,更改为HP000001并输入该值,然后再次按下Utility按钮。 然后它将“不发表”。 通知这一点,然后执行cal,然后将安全代码设置回默认值。 以上来自于谷歌翻译 以下为原文 This procedure is poorly documented and I will get it fixed. You have to put the display in the mode that displays HP000000 or AT000000. Then short the rectangular pads, change to HP000001 and enter that value and press the Utility button again. It will then be "unsecurred". Veryify this, then do the cal, and then set the security code back to the default. |