我刚刚在我的N9912A上安装了最新的固件(A.06.02)。
我正在观察频谱图显示,发现可能代表硬件故障的东西。 我在显示屏的右下角看到一块总共是黑色的8像素(4宽2高)。 我之前从未注意到这一点,并在固件升级后立即看到它,所以我很确定它刚刚发生。 经过一个小时左右的CAT,NA和SA模式,我发现在显示器中有扫描速度标签的所有情况下像素都“死”,而死像素总是在 相对于扫描速度标签的相同位置:距标签基线约1.25厘米,并且距显示器右边缘约2.5毫米。 即使在“分屏”频谱图显示中,死像素也始终遵循扫描速度标签。 我附加了三个屏幕截图来说明这些观察结果,第三个屏幕截图显示全屏模式下没有坏点(当显示屏上没有扫描速度标签时)。 为莫尔条纹道歉,但你会明白这一点。 这是显示内存故障还是固件错误? 我没有尝试降级回A.05.53以查看问题是否存在,但这可能是下一步。 斯坦 以上来自于谷歌翻译 以下为原文 I just installed the latest firmware (A.06.02) on my N9912A. I was observing a spectrogram display and noticed something that may represent a hardware failure. I saw a block of eight pixels (4 wide by 2 high) in the lower right corner of the display that were always black. I had never noticed this before, and saw it right away after the firmware upgrade, so I'm pretty sure it just recently happened. After an hour or so of playing around with the CAT, NA, and SA modes, I find that the pixels are "dead" in all those situations where there is a sweep speed label in the display, and the dead pixels are always in exactly the same position relative to the sweep speed label: about 1.25 cm above the baseline of the label, and inset about 2.5 mm from the right edge of the display. Even in the "split screen" spectrogram display, the dead pixels always follow the sweep speed label. I've attached three screen shots that illustrate these observations, the third one showing no dead pixels in the full screen mode (when there is no sweep speed label on the display). Apologies for the moiré patterns, but you'll get the idea. So is this a display memory failure or a firmware bug? I haven't tried downgrading back to A.05.53 to see if the problem follows, but that may be the next step. Stan |
|
相关推荐
4个回答
|
|
|
|
|
|
Stan,有一个像素测试SCPI命令可用于测试(子)像素活动。 将以下命令发送到仪器,并报告这些感兴趣的像素是否亮起://全屏BLACK DIAG:SCOL 0 //全屏蓝色DIAG:SCOL #hFF //全屏绿色DIAG:SCOL#hFF00 //全屏RED DIAG:SCOL #hFF0000 //正常显示DIAG:SCOL -1请注意,不支持DIAGnostic SCPI命令,并且可能随时更改。 请不要写任何软件反对他们! Wil编辑:wils于2012年8月13日10:28 AM哦,我明白了。 在全屏模式下看起来像素是正常的。 所以上面的测试可能不会表明任何新的东西。 我们将不得不检查固件错误...编辑:wils于2012年8月13日上午10:47已在内部验证该问题不是硬件像素问题。 错误提交。 以上来自于谷歌翻译 以下为原文 Stan, There's a pixel testing SCPI command that you can use to test for (sub)pixel activity. Send the following commands to you instrument, and report back if those pixels of interest light up or not: // Full screen BLACK DIAG:SCOL 0 // Full screen BLUE DIAG:SCOL #hFF // Full screen GREEN DIAG:SCOL #hFF00 // Full screen RED DIAG:SCOL #hFF0000 // Normal Display DIAG:SCOL -1 Please note that DIAGnostic SCPI commands are not supported, and are subject to change at any time. Please do not write any software against them! Wil Edited by: wils on Aug 13, 2012 10:28 AM Oh, I see. It looks like the pixels are OK as seen in the full screen mode. So the above test probably won't indicate anything new. We'll have to check for a firmware bug... Edited by: wils on Aug 13, 2012 10:47 AM Verified the issue in-house as NOT a hardware pixel issue. Bug submitted. |
|
|
|
zyh34997 发表于 2019-5-23 09:55 此缺陷已在内部版本中修补,并将在某些尚未确定的时间发布。 会 以上来自于谷歌翻译 以下为原文 This defect has been patched in an internal build, and will be released at some yet to be determined time. Wil |
|
|
|
谢谢! 改变频谱图和瀑布显示的“全屏”模式的行为有什么进展吗? 如果通过“光谱图设置”软键调用全屏模式,它不会使图像变大,它只会消除光谱轨迹以及上方,下方和左侧的标签。 谱图图像。 (“瀑布设置”菜单中没有“全屏”命令。)如果通过“系统”菜单调用频谱图或瀑布的全屏命令,则行为不同:频谱图或瀑布分量 显示器保持相同的大小,同时显示器的频谱分量扩展到全屏,并且网格线被展开和显示,但仅在屏幕中没有光谱图或瀑布的部分。 此致,斯坦 以上来自于谷歌翻译 以下为原文 Thanks! Any progress on changing the behavior of the "full screen" mode for the spectrogram and waterfall displays? If you invoke the full screen mode through the "Spectrogram Settings" soft key, it doesn't make the image any larger, all it does is to blank out the spectrum trace and the labels above, below, and to the immediate left of the spectrogram image. (There is no "Full Screen" command in the "Waterfall Settings" menu.) If you invoke the full screen command for either the spectrogram or waterfall through the "System" menu, the behavior is different: The spectrogram or waterfall component of the display remains the same size, while the spectrum component of the display expands to full screen, and the grid lines are expanded and displayed, but only in the portion of the screen where the spectrogram or waterfall isn't. Regards, Stan |
|
|
|
只有小组成员才能发言,加入小组>>
1238 浏览 0 评论
2352 浏览 1 评论
2163 浏览 1 评论
2029 浏览 5 评论
2913 浏览 3 评论
983浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
713浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
813浏览 0评论
1239浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-11-28 23:30 , Processed in 1.317367 second(s), Total 55, Slave 48 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号