嗨,我正在使用网络分析仪:E5071C我在查询SCPI命令(使用labview)时将功率电平归零,同时尝试使用cal方法作为响应(Thru)手动相同的事情将设置如下:Cal >> Calibrate>
>响应(直通)>> SelectPorts(2,1)>>通过>>完成我试图使用keysight网站提供的帮助找到相应的命令:http://ena.support.keysight.com/e5071c/manuals/ webhelp / eng /网络anaylzer上的软键:等价SCPI选择端口:SENS1:CORR:COLL:THRU 2,1通过:SENS1:CORR:COLL:THRU 2,1; * OPC? 完成:SENS1:CORR:COLL:SAVE“我将配置所需的配置如下:CALC1:PAR1:DEF S21:SENS1:SWE:TYPE POW:SOUR1:POW:STAR -45:SOUR1:POW:STOP 5:SENS1: FREQ 2620000000:SENS1:SWE:POIN 201:SENS1:BAND 1E3然后尝试运行命令:SENS1:CORR:COLL:THRU 2,1出错:*需要额外的标准。*请帮我找到正确的方法 使用LabVIEW或SCPI命令集执行Cal >> Resp(thru)方法。提前谢谢!关于Geeta Kannammanavar 以上来自于谷歌翻译 以下为原文 Hi, I am using Network analyzer: E5071C I have query on SCPI command(using LabVIEW) to zeroing the power level while trying to use the cal method as response(Thru) Manually same thing will be set as below: Cal>>Calibrate>>Response(Thru)>>SelectPorts(2,1)>>Thru>>Done I tried to find the equivalent command using the help available in keysight website: http://ena.support.keysight.com/e5071c/manuals/webhelp/eng/ Soft Keys on Network anaylzer: Equicvalent SCPI Select Ports :SENS1:CORR:COLL:THRU 2,1 Thru :SENS1:CORR:COLL:THRU 2,1;*OPC? Done :SENS1:CORR:COLL:SAVE" I will configure the required config as below :CALC1:PAR1:DEF S21 :SENS1:SWE:TYPE POW :SOUR1:POW:STAR -45 :SOUR1:POW:STOP 5 :SENS1:FREQ 2620000000 :SENS1:SWE:POIN 201 :SENS1:BAND 1E3 Then try to run the command :SENS1:CORR:COLL:THRU 2,1 Getting an error : *Additional standard needed.* Please help me to find the right way to execute the Cal>>Resp(thru) method using LabVIEW or SCPI command set. Thank You in advance!. Regards Geeta Kannammanavar |
|
相关推荐
2个回答
|
|
亲爱的Geeta,错误“需要附加标准”表明,在完成计算校准系数所需的所有数据测量之前,已经发送了打开校准功能的GPIB命令。
如果你想通过校准执行响应,请尝试定义如下,SENS1:CORR:COLL:METH:THRU 1,2 SENS1:CORR:COLL:THRU 1,2 * OPC? SENS1:CORR:COLL:SAVE 以上来自于谷歌翻译 以下为原文 Dear Geeta, The error "Additional standard needed" shows that the GPIB command that turns ON the calibration function has been sent before all the data measurements required to calculate the calibration factor have been completed. if you want to perform response through calibration, please try to define as follows, SENS1:CORR:COLL:METH:THRU 1,2 SENS1:CORR:COLL:THRU 1,2 *OPC? SENS1:CORR:COLL:SAVE |
|
|
|
大洼球王 发表于 2018-11-6 20:28 谢谢Takeshio! 以上来自于谷歌翻译 以下为原文 Thank You Takeshio ! |
|
|
|
只有小组成员才能发言,加入小组>>
1271 浏览 0 评论
2371 浏览 1 评论
2187 浏览 1 评论
2061 浏览 5 评论
2943 浏览 3 评论
1085浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
744浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
906浏览 0评论
1271浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-19 02:41 , Processed in 1.296959 second(s), Total 81, Slave 64 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号