电子发烧友论坛|
你好。
在将DSO-X 2012A更新到最新固件后,我发现了一个相当奇怪的问题,不确定它是由WFG造成的,还是由于我身边缺乏咖啡。 使用500欧姆电缆将WFG-OUT连接到CH1-IN,将WFG设置为1kHz方波,Hi-Pot = 1V,Lo-Pot = 0V,阻抗= 500欧姆。 在测量信号时,示波器向我显示2Vpp的信号,而不是预期的1Vpp。 连接显示测量信号,仍显示WFG设置。 所以问题出现了,如果我宁愿上床睡觉并称它为一天,还是应该这样? 以上来自于谷歌翻译 以下为原文 Hello. After updating the DSO-X 2012A to the latest firmware, i found a rather curious problem, unsure if it's caused by the WFG, or by a lack of coffee on my side. Connected the WFG-OUT to the CH1-IN using a 50Ohms cable, set the WFG to 1kHz square wave, Hi-Pot=1V, Lo-Pot=0V, Impedance=50Ohms. On measuring the signal, the scope shows me a signal with 2Vpp, instead of the 1Vpp as expected. Attached the display of the measured signal, with WFG settings still shown. So the question arises, should i rather go to bed and call it a day, or is this supposed to be this way? 附件
4个回答
将WAVGEN设置为50欧姆时,您希望将1Vpp放入50欧姆负载。
所以它的开路电压是2Vpp,这是你看到CH1设置为1Meg输入阻抗。 设置CH1 - >阻抗 - > 50欧姆。 然后它将显示1Vpp到50欧姆的示波器阻抗。 以上来自于谷歌翻译 以下为原文 With the WAVGEN set to 50 ohms it expects you want 1Vpp into a 50 ohm load. So its open circuit voltage is 2Vpp which is what you are seeing with CH1 set to 1Meg input impedance. Set CH1-->Impedance-->50ohms. Then it will show 1Vpp into the 50 ohm scope impedance. |
dfgsgdg 发表于 2019-4-25 19:48 啊,所以这个设备确实让我心烦意乱。 谢谢你解决我的神秘面纱。 我唯一无法工作的是改变通道的阻抗。 并不重要,因为我现在知道将WFG留在High-Z。 但我在通道菜单中找不到任何改变输入阻抗的选项。 通道菜单中没有标有“阻抗”的软按钮。 如果你能给我另一个推动,那么我确实找到了这个选项,我将非常感激。 提前再次感谢您的帮助。 以上来自于谷歌翻译 以下为原文 Ahhh, so the device did outsmart me for good. Thank you for solving that mystery of mine. The only thing i couldn't quite get to work was changing the impedance of the channel. Not that it would matter, as i now know to leave the WFG at High-Z. But i couldn't find any option in the channel menu for changing the input impedance. There is no soft button labelled with "Impedance" in the channel menu. If you could give me another nudge, so i do find that option, i'd be very grateful. Thank you again, in advance, for your help. |
2000系列示波器的输入阻抗固定为1M。 它可以在High-Z和50欧姆之间切换,适用于3000和更高系列的示波器。 从90000系列示波器开始,它固定为50欧姆。 人 以上来自于谷歌翻译 以下为原文 The input impedance on the 2000 series of scopes is fixed at 1M. It is switchable, between High-Z and 50 Ohms, on the 3000 and higher families of scopes. Starting with the 90000 series of scopes, it is fixed at 50 Ohms. Al |
nvywyerwer 发表于 2019-4-25 20:05 这可以解释某些缺乏改变输入阻抗的选择。 非常感谢真正快速的回复。 以上来自于谷歌翻译 以下为原文 That would explain the certain lack of the option to change the input impedance. Thanks a lot for the really fast reply. |
只有小组成员才能发言,加入小组>>
290 浏览 0 评论
1871 浏览 0 评论
2757 浏览 1 评论
2658 浏览 1 评论
2478 浏览 5 评论
1871浏览 0评论
434浏览 0评论
290浏览 0评论
/9
GMT+8, 2025-12-14 16:24 , Processed in 0.804554 second(s), Total 78, Slave 61 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
淘帖
2012