嗨,我让我的34972A通信了一段时间,但现在安捷伦连接专家(ACE)表示“存在通信失败”并且错误是“仪器未能返回IDN字符串”。
我无法再通过我的代码或Agilent Interactive I / O与仪器通信。 我已经卸载并重新安装了ACE ...当第一次重新安装ACE时,它表明还有其他VISA驱动程序。 我卸载了所有我看到的(NI和TEK),然后重新安装了ACE ...仍然没有...任何建议? 谢谢 以上来自于谷歌翻译 以下为原文 Hi, I had my 34972A communicating awhile back but now the Agilent Connection Expert (ACE) indicates "there was a commuincation failrue" and the error is "The instrument failed to return an IDN string". I can no longer communicate with the instrument through my code or Agilent Interactive I/O. I have uninstalled and reinstalled ACE... when reinstalling ACE the first time it indicated there were other VISA drivers. I uninstalled all that I coudl see (NI and TEK) and then reinstalled ACE... still no go... any suggestions? Thanks 附件
|
|
相关推荐
2个回答
|
|
以下是我要做的事情:1。从以下2个目录中删除visa32.dll文件:c: windows system32 c: windows syswow64 2.完全卸载Agilent IO Libraries Suite 3.下载最新版本的
来自[www.agilent.com/find/iosuite]的IO库并重新安装 以上来自于谷歌翻译 以下为原文 Here is what I would do: 1. Delete the visa32.dll file from the following 2 directories: c:windowssystem32 c:windowssyswow64 2. Completely uninstall the Agilent IO Libraries Suite 3. Download the latest version of the IO Libraries from [www.agilent.com/find/iosuite] and reinstall |
|
|
|
仪器可能具有错误的地址设置,或者两个或多个仪器设置为相同的地址。
Connection Expert中配置的地址必须与仪器的地址设置相匹配。 您可以从前面板检查34972A地址。 由于它正在工作,为34972A循环供电并刷新ACE应该重新建立与34972A的通信。 我将与IO Libraries专家核实并回复您。 以上来自于谷歌翻译 以下为原文 The instrument may have the wrong address setting or two or more instruments are set to the same address. The address configured in Connection Expert must match the address setting of the instrument. You can check the 34972A address from the front panel. Since it was working, cycling power to the 34972A and refreshing ACE should reestablish communication with the 34972A. I will check with the IO Libraries expert and get back to you on this. |
|
|
|
只有小组成员才能发言,加入小组>>
1287 浏览 0 评论
2375 浏览 1 评论
2194 浏览 1 评论
2064 浏览 5 评论
2951 浏览 3 评论
1111浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
753浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
926浏览 0评论
1287浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-24 03:12 , Processed in 1.610605 second(s), Total 48, Slave 42 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号