电子发烧友论坛|
嗨,我们有一个E4436B ESG,最近当我们尝试进行IQ校准时,报告了两个错误:517校准失败:ARB Cal 517校准失败:偏移校准这些错误是否表明内部硬件出现故障?
如果没有,那可能对他们有什么解决方法? 谢谢。 以上来自于谷歌翻译 以下为原文 Hi, We have a E4436B ESG and recently when we tried to do a IQ calibration, two errors were reported: 517 Calibration failure: ARB Cal 517 Calibration failure: Offset Cal Do these errors suggest internal hardware malfunction? If not, what might be the possible fix for them? Thank you. |
3个回答
嗨 - 我和你的E4436B支持工程师谈过。
她的回答是,“这表明硬件故障,无论是基带还是输出板。都需要服务。” 问候 - 以上来自于谷歌翻译 以下为原文 Hi - I talked to my E4436B support engineer for you. Her reply was, "This would indicate hardware failure, either Baseband or output board. It would require service." Regards - |
嗨tabbott,谢谢你的回复。 在frac-N模块中,单元可能有一个坏芯片(1GC1-4210),我认为这部分会导致IQ校准失败。 我确实在1GC1-4210上有另一个问题。 该芯片似乎声名狼借,因为它经常变坏。 1GC1-4210的问题与坏批次有关吗? 什么日期代码可以被认为是“好”? 谢谢。 以上来自于谷歌翻译 以下为原文 Hi tabbott, Thanks for your reply. The very unit may have a bad chip (1GC1-4210) in frac-N module and I think that partly contributes to failure of IQ calibration. I do have another question on 1GC1-4210 though. The chip seems to have a bad reputation as it goes bad quite often. Is the problem with 1GC1-4210 associated with a bad batch? After what date code can be assumed to be "good"? Thank you. |
60user155 发表于 2018-11-23 20:50 Hi - 缺陷1GC1-4210有HP标记,良好1GC1-4210有Agilent标记。 有缺陷的芯片将给出错误511“输出不均匀”,并导致低于1GHz的无输出或不稳定输出。 我不认为它会对IQ cal失败产生影响。 问候 - 以上来自于谷歌翻译 以下为原文 Hi - Defective 1GC1-4210 have HP markings, good 1GC1-4210 have Agilent markings. The defective chip will give Error 511 "Output unleveled" and causes NO OUTPUT or UNSTABLE OUTPUT below 1GHz. I don’t think it would have an effect on the IQ cal failure. Regards - |
只有小组成员才能发言,加入小组>>
1844 浏览 0 评论
2739 浏览 1 评论
2640 浏览 1 评论
2449 浏览 5 评论
3458 浏览 3 评论
1845浏览 0评论
417浏览 0评论
/9
GMT+8, 2025-12-2 12:14 , Processed in 0.728115 second(s), Total 76, Slave 59 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
淘帖
921