我有一个命令发送到我的n5242a,SENS:路径:CONF:SEL“ONT_PATH”但它显示错误“无效的配置路径”,我找不到编程指南的这个命令集部件号:E8356-90028。
我想知道这个ONT_PATH实际上是设备中的文件夹或只是一个参考。 以上来自于谷歌翻译 以下为原文 I have a command to send to my n5242a, SENS:PATH:CONF:SEL "ONT_PATH" but it showed error "invalid configuration path", i can not find this command set from programming guide Part Number: E8356-90028. I want to know this ONT_PATH is actual a folder in the equipment or just a reference. |
|
相关推荐
2个回答
|
|
嗨,我可以使用此命令调用之前在N5242A上保存的路径配置,这是我的步骤:1。菜单跟踪/通道 - >通道 - >硬件设置 - >路径配置...打开路径配置对话框
,对路径配置进行一些更改(例如,如果N5242A具有内部组合器,则将端口1旁路开关更改为组合路径),然后单击“存储...”按钮以保存名称为“ONT_PATH”的路径配置,单击“确定”。 2.预设PNA-X以将路径配置恢复为默认值3.使用Agilent Connection Expert设置仪器连接,向N5242A发送命令:SENS:PATH:CONF:SEL“ONT_PATH”4。立即检查路径配置, 应该选择“ONT_PATH”有几件事要检查:a。 也许您的固件版本不支持此命令,您可以尝试将固件升级到最新的A.09.90.13并尝试b。 确保已成功存储路径配置,可以使用命令SENSe:PATH:CONFig:CATalog? 验证是否有名称为“ONT_PATH”的配置c。 您是否使用C#或C ++等编程语言来控制仪器? 为了避免“问题”,你可以尝试SENS:PATH:CONF:SEL'ONT_PATH'顺便说一下,你可以从http://na.support.keysight.com/pna/help下载chm格式的PNA帮助文件 /并查找编程部分下的命令 以上来自于谷歌翻译 以下为原文 Hi, I am able to use this command to recall the path configuration saved before on N5242A, here are my steps: 1. Menu Trace/Channel->Channel->Hardware Setup->Path Config... to open the Path configuration dialog, make some changes on the path configuration (such as changing the Port 1 bypass switch to Combiner Path if your N5242A has internal combiner), then click "Store..." button to save the path configuration with name "ONT_PATH", click OK 2. Preset the PNA-X to restore the path configuration to default 3. Set up a instrument connection with Agilent Connection Expert, send command to the N5242A: SENS:PATH:CONF:SEL "ONT_PATH" 4. Check the path configuration now, the "ONT_PATH" should be selected There are several things to check: a. Maybe your firmware revision doesn't support this command, you can try to upgrade your firmware to the latest A.09.90.13 and have a try b. Make sure you have stored the path configuration successfully, you can use command SENSe:PATH:CONFig:CATalog? to verify if there is a config with name "ONT_PATH" c. Are you using any programming language such as C# or C++ to control the instrument? to avoid any problem with the ", you can try SENS:PATH:CONF:SEL 'ONT_PATH' By the way, you can download the chm format PNA help file from http://na.support.keysight.com/pna/help/ and look for the commands under Programming part |
|
|
|
|
|
|
|
只有小组成员才能发言,加入小组>>
1287 浏览 0 评论
2375 浏览 1 评论
2194 浏览 1 评论
2064 浏览 5 评论
2951 浏览 3 评论
1111浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
753浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
926浏览 0评论
1287浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-23 23:22 , Processed in 1.478961 second(s), Total 78, Slave 62 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号