我已经使用我的'MSO-X 3012A一个星期左右没有问题,并且对目前的表现非常满意。
今天我想我可以安装演示许可证选项以查看串行解码功能,但在安装30天演示许可证后,我发现我有一些非常奇怪的问题,这些问题以前没有。 1.在任何触发设置选项中,通道1不再可用作触发源。 在安装30天试用许可证之前一切正常,所以我的猜测是由于下拉菜单中触发器选项的数量增加可能超出了显示能力。 这个理论似乎通过使用BenchView程序得到证实,在“仪器设置”选项卡中我将通道1作为触发选项。 2.选择USB作为触发类型,D +或D-的选项使通道1变灰。 通道2可用,如D1-12等。这可能与上述问题有关? 我附上了照片,证明了这一点,因为我无法使用屏幕捕获工具来显示打开的菜单,因为我已经能够确定。 有没有人有任何想法如何解决这个问题? 谢谢,Chris IMG_0584.JPG264.1 KB 以上来自于谷歌翻译 以下为原文 I have been using my 'MSO-X 3012A for a week or so with no problems and very pleased with the performance so far. Today I thought I might install the Demo Licence option to have a look at the serial decode functionality but after installing the 30 Day demo licence I find that I have a couple of very odd problems which weren't there previously. 1. Channel 1 is no longer available as a trigger source in any trigger setting option. This had been fine before the 30-day trial licence was installed so my guess is that it is due to the increased number of trigger options in the drop down menu perhaps exceeding the display ability. This theory seems to be confirmed by using the BenchView program where in the Instrument Settings tab I have Channel 1 as a trigger option. 2. Selecting USB as a trigger type, the options for D+ or D- have Channel 1 grayed out. Channel 2 is available as are the D1-12 etc. This is possibly connected to the problem above? I have attached photo's that demonstrate this as it isn't possible to use the screen capture facility to display open menu's as far as I have been able to ascertain. Does anyone have any ideas as to how to fix this problem? Thanks, Chris 附件
|
|
相关推荐
2个回答
|
|
嗨Chris,如果双阈值串行解码(ARINC 429或MIL-STD-1553)处于活动状态,则它们的源不能用于单阈值硬件触发。
我们的触发电路可以针对给定通道以单阈值或双阈值模式进行编程,但不能同时对两者进行编程。 您的屏幕截图中的活动串行解码是否都是这些模式? 如果是这样,请尝试关闭串行解码或切换到其他解码模式,并查看通道1是否作为可选触发源返回。 谢谢, - 克里斯 以上来自于谷歌翻译 以下为原文 Hi Chris, If dual-threshold serial decodes (ARINC 429 or MIL-STD-1553) are active, their source(s) cannot be used for single-threshold hardware triggers. Our trigger circuitry can be programmed in a single- or dual-threshold mode for a given channel, but not both simultaneously. Is the active serial decode in your screenshots either of these modes? If so, try turning off serial decode or switching to a different decode mode and see if channel 1 returns as a selectable trigger source. Thanks, --Chris |
|
|
|
小新1999 发表于 2018-10-10 15:47 嗨,克里斯,我实际上是在触发USB。 我将Ch1连接到D +线,将Ch2连接到D-线。 在进行Demo安装之前,我一直在对USB触发进行一些测试,所以我知道它没问题。 当我安装Demo时,它甚至还连接到USB分线盒,因此我很惊讶这发生了。 正如您在第一张图片中看到的那样,当尝试选择D +源通道时,实际上是灰色的,而在图2中,当我尝试将其设置为触发源时,我需要在SOP上触发,通道1不是 甚至出现在源下拉框中。 至于串行解码,我没有在那里设置任何东西并且没有选择串行按钮。 问候,克里斯 以上来自于谷歌翻译 以下为原文 Hi Chris, I was actually triggering on USB. I had Ch1 connected to the D+ line and Ch2 to the D- line. I had been doing some testing on USB triggering before doing the Demo install so I know that it was OK. It was even still connected up to the USB break-out box while I installed the Demo's hence my surprise that this had happened. As you can see in the first picture when trying to select the D+ source channel 1 is actually greyed out, and in picture 2 when attempting to set it as the trigger source, which I needed to do to trigger on SOP, channel 1 isn't even showing up in the source drop down box. As for serial decodes, I hadn't got around to setting anything up in there and had not selected the Serial button. Regards, Chris |
|
|
|
只有小组成员才能发言,加入小组>>
1230 浏览 0 评论
2351 浏览 1 评论
2160 浏览 1 评论
2026 浏览 5 评论
2908 浏览 3 评论
974浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
707浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
808浏览 0评论
1230浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-11-26 09:11 , Processed in 2.476934 second(s), Total 49, Slave 42 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号