You can't, not really.
The front end of the scope has a BW of at least 600MHz. If you just slow down the sample rate, then the BW gets lower, but without a HW anti-aliasing flter, all of the higher frequency signals will show up, just aliased into your lower frequency range. You could make this better by adding an external HW filter, but it would have to be a really good filter to work well.
If you are using a probe, or have the channels set for 1Mohm impedance, you can turn on the built-in 20MHz filter which will get you closer.
You can also use the "Low-Pass Filter" function, but I don't know what the pass-band/stop-band characteristics are at that low a frequency.
If that's the frequency range you want to work in, you are probably using the wrong scope.
Al
Disclaimer: For more reliable response, you should consider calling your local Agilent Technical Call Center. The Agilent Forums are monitored on an "as available" basis, and aren't necessarily the fastest way to get technical questions answered.
你不能,不是真的。
示波器的前端具有至少600MHz的BW。
如果您只是降低采样率,那么BW会降低,但如果没有HW抗混叠滤波器,所有较高频率的信号都会显示出来,只是混叠到较低的频率范围内。
您可以通过添加外部硬件过滤器来改善这一点,但它必须是一个非常好的过滤器才能正常工作。
如果您正在使用探头,或者将通道设置为1Mohm阻抗,您可以打开内置的20MHz滤波器,这将使您更接近。
您也可以使用“低通滤波器”功能,但我不知道在低频率下通带/阻带特性是什么。
如果这是您想要使用的频率范围,则可能使用了错误的范围。
免责声明:为了获得更可靠的响应,您应该考虑致电当地的安捷伦技术呼叫中心。
安捷伦论坛在“可用”的基础上进行监控,并不一定是解决技术问题的最快方式。
以上来自于谷歌翻译
以下为原文
You can't, not really.
The front end of the scope has a BW of at least 600MHz. If you just slow down the sample rate, then the BW gets lower, but without a HW anti-aliasing flter, all of the higher frequency signals will show up, just aliased into your lower frequency range. You could make this better by adding an external HW filter, but it would have to be a really good filter to work well.
If you are using a probe, or have the channels set for 1Mohm impedance, you can turn on the built-in 20MHz filter which will get you closer.
You can also use the "Low-Pass Filter" function, but I don't know what the pass-band/stop-band characteristics are at that low a frequency.
If that's the frequency range you want to work in, you are probably using the wrong scope.