我正在通过同轴到波导转换直接驱动PNA-X上的波导设置。
这是一个单端口测量,我的DUT连接到波导。 我想通过使用我的ECal单元来纠正因电缆运动引起的系统漂移,从而提高稳定性。 我基本上希望对待我的波导系统,就像85070E软件一样对待探头并随着时间的推移刷新ECal校准。 单独使用PNA单元是否可行,或者我是否需要一些PNA中没有直接提供但已写入85070E软件的专用脚本? 我已经为波导单元设置了校准套件,其具有短的偏移短路和负载作为标准。 以上来自于谷歌翻译 以下为原文 I am driving a waveguide setup directly off the PNA-X via a coax-to-waveguide transition. This is a one-port measurement with my DUT attached to the waveguide. I would like to improve stability by using my ECal unit to correct for system drift caused by cable motion. I'm basically looking to treat my waveguide system like the 85070E software treats the probe and refresh the ECal calibration over time. Is this possible with a PNA unit alone, or do I need some specialized scripting that is not available directly in the PNA but has been written into the 85070E software? I already have cal kit for the waveguide unit set up with a short, offset short, and a load as standards. |
|
相关推荐
3个回答
|
|
谢谢,戴夫。
我已经潜伏了很多年 - 在过去的9年里,我的实验室里有一个PNA,现在还有一个PNA-X。 最后迫切需要我认为论坛会有所帮助。 以下是我对85070E / ECal us的理解 - 85070E探头软件基本上进行了两次校准 - 一次用于ECal,另一次用于探头。 探头使用空气/短路/水进行校准,ECal使用其内部标准。 在每次测量之前,探头软件可以刷新电缆的校准,从而消除电缆相位的变化,但保留探头的响应。 从本质上讲,它是某种形式的港口扩张或固定? 杰夫 以上来自于谷歌翻译 以下为原文 Thanks, Dave. I've been lurking for many years - I've had a PNA in my lab the last ~ 9 years, and have a PNA-X as well now. Finally had an urgent need that I thought the forum would be helpful for. Here's my understanding of the 85070E / ECal us - the 85070E probe software essentially does two calibrations - one for the ECal and one for the probe. The probe uses air/short/water for its calibration, and the ECal uses its internal standards. Before each measurement, the probe software can refresh the calibration of the cable, which removes changes in the phase of the cable, but retains the response of the probe. Essentially, it's some form of port extention or fixturing? Jeff |
|
|
|
理论上 - 您可以将ECal留在电缆的同轴端和同轴电缆到波导转换之间,如果您以足够低的功率运行(低于ECal压缩点),并且您愿意编写一大堆软件 要进行重新校正数学运算,您可以设置一个系统,在该系统中,您可以针对校准后发生的任何温度或运动漂移(幅度和相位)校正测量值。 但是,这不是我推荐的DIY项目。 但好消息是我们已经为您完成了这项工作,并创建了一个类似Ecal的模块,该模块设计为信号路径中的机构,并附带软件,可以为您完成所有重新校正数学运算。 我们将其称为CalPod,您可以找到有关这些模块及其功能的更多信息[_ * HERE * _ | www.home.agilent.com/en/pc-2274317/calpod-calibration-refresh-module?nid = -33010.0 0.00&安培; CC = US&安培; LC = ENG] 以上来自于谷歌翻译 以下为原文 theoretically - you could leave the ECal in place between the coax end of your cable and the coax to waveguide transition and if you are operating at low enough power (below the ECal compression point), and you are willing to write a whole bunch of software to do the re-correction math, you could setup a system where your measurements are corrected for any temperature or movement drift (magnitude and phase) that has occurred after calibration. However, that is not what I would recommend as a DIY project. But the good news is that we have done this work for you and created an Ecal-like module that is designed to be institu in the signal path and comes with the software that will do all the re-correction math for you. We call it the CalPod and you can find more information about these modules and their capabilities [_*HERE*_|www.home.agilent.com/en/pc-2274317/calpod-calibration-refresh-module?nid=-33010.0.00&cc=US&lc=eng] |
|
|
|
Topcbpcba 发表于 2019-3-1 14:31 达拉斯是对的。 这样做非常困扰,Calpod为您完成所有工作,温度补偿范围为-30至+ 80C。 要做你想做的事情,你可以通过将你的ecal直接放入,WG适配器直接连接到它来做到这一点。 然后,在WG端口校准系统; 它需要是一个2端口cal,所以你需要第二个WG适配器转到端口2.然后,用WG端口表征ecal,但奇怪的是端口1位于Ecal的远端。 现在,当您执行1端口Ecal时,它是指连接Ecal远端端口的WG连接器的输出。 重新校准,然后测量。 Calpod做同样的事情,体积小得多,成本约为1/2(不包括控制器)。 以上来自于谷歌翻译 以下为原文 daras is right. It is a lot of fussing around to do that, and Calpod does it all for you, with temperature compensation from -30 to +80C. To do what you what to do, you could do it pretty simply by putting your ecal in-line with, with the WG adapter connected directly to it. Then, calibrate the system at the WG port; it will need to be a 2 port cal, so you need a second WG adapter to go to port 2. Then, characterize the ecal with WG port, but the odd thing is the port 1 is on the far side of the Ecal. Now, when you do a 1 port Ecal, it is reference to the output of the WG connector connected the far-end port of the Ecal. Recalibrate, then measure. Calpod does the same thing, is a lot smaller, and costs about 1/2 as much (not counting the controller). |
|
|
|
只有小组成员才能发言,加入小组>>
1221 浏览 0 评论
2346 浏览 1 评论
2155 浏览 1 评论
2021 浏览 5 评论
2900 浏览 3 评论
964浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
700浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
801浏览 0评论
1223浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-11-22 14:08 , Processed in 1.810663 second(s), Total 81, Slave 64 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号