您好,我的3458A,年份2001,固件8,2,最近得到了NVRAM的组件级交换。
现在我想知道,如果在安捷伦商店中提供的固件9,x,修复了任何进一步的错误,或固件8和9之间有什么区别? 没有可用的文档。 请问有人解释一下吗? 问FrankEdited:Frank博士于2013年12月13日上午6:22 以上来自于谷歌翻译 以下为原文 Hello, my 3458A, vintage 2001, firmware 8,2, recently got a component level exchange of NVRAMs. Now I wonder, if firmware 9,x available in the agilent store, fixes any further bugs, or what is the difference between firmware 8 and 9? There's no documentation available. Would anybody explain that, please? regards Frank Edited by: Dr. Frank on Dec 13, 2013 6:22 AM |
|
相关推荐
3个回答
|
|
嗨,弗兰克博士,从更新版本8,2到版本9,2更新中没有“用户影响”更改,这些更改偶然提到:9 - 仪器固件版本2 - A3 / A13组件固件版本作为参考,我有
附上一份包含3458A修订历史的文件。 以上来自于谷歌翻译 以下为原文 Hi Dr. Frank, There were no 'user impacted' changes made in the update from revison 8,2 to revision 9,2 which incidentally refers to: 9 - instrument firmware revision 2 - A3/A13 assembly firmware revision For reference, I have attached a document containing the revision history of the 3458A. 附件
|
|
|
|
子虚道人 发表于 2018-11-21 17:25 你好jdwindrvr,谢谢你的暗示。 我已经知道3458A的服务说明,但仍然缺乏对固件#9的描述。 由于此版本是在HP重命名为agilent时发布的,因此我认为FIB 9也会重命名GPIB消息,但事实并非如此。 ID? 仍然提供“HP3458A”(在更新的仪器上找到)。 所以问题仍然存在,这个最新的FW版本的内容是什么,或者有什么好处? FrankEdited:Frank博士于2013年12月14日凌晨4点46分 以上来自于谷歌翻译 以下为原文 Hello jdwindrvr, thank you for your hint. I already know the Service Notes for the 3458A, but a description of firmware#9 still is lacking. As this version was published when HP was renamed to agilent, I assumed that the GPIB messages would be renamed by FW 9 also, but that's not the case. ID? still delivers "HP3458A" (as found on an updated instrument). So the question remains, what's the content, or what's the benefit of this latest FW version? Frank Edited by: Dr. Frank on Dec 14, 2013 4:46 AM |
|
|
|
dmm156 发表于 2018-11-21 17:36 嗯,有可能以不同的方式解决这个奇迹。 显然,384kB中只有10个字节在vers之间是不同的。 除了校验和字节,版本计数,实际上只有4个字节已经改变,这改变了一个或两个限制参数(整数),这些参数可能在内部测试中使用。 这解释了,为什么没有这种变化的文档,因为没有真正的错误修复或改进。 显然没有必要将版本8,x仪器更新到最新版本。 坦率 以上来自于谷歌翻译 以下为原文 Well, it was possible to solve that miracle in a different way. Obviously, only 10 bytes of the 384kB are different between vers. 8 and 9. Besides checksum bytes, the version count, only 4 bytes really had been changed, which alter one or two limit parameters (integer numbers), which are presumably used in internal testing. That explains, why there is no documentation of this change, as there is no real bug fixing or improvement. It's obviously not necessary at all to update a version 8,x instrument to the latest revision. Frank |
|
|
|
只有小组成员才能发言,加入小组>>
1283 浏览 0 评论
2371 浏览 1 评论
2189 浏览 1 评论
2062 浏览 5 评论
2946 浏览 3 评论
1105浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
752浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
923浏览 0评论
1285浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-23 03:36 , Processed in 2.061476 second(s), Total 80, Slave 64 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号