大家好,因为我们在实验室中有很多旧的安捷伦设备,我们习惯通过Agilent IO库套件与它们进行通信。
我刚试过BenchVue来接触我们新的N9020A MXA频谱分析仪(我想知道最近的N5171B是否兼容)。 通过Agilent IO连接专家可以访问设备,但是在BenchVue上仍然可以看到红叉,我无法处理它。 ping和RDP连接运行良好。 有没有人知道我可以改变什么参数来强制BenchVue接口上的连接? 非常感谢 ! 以上来自于谷歌翻译 以下为原文 Hi all, as we have a lot of old Agilent devices here in the lab, we were used to communicate with them by means of the Agilent IO libraries suite. I just have tried BenchVue to reach our new N9020A MXA spectrum analyzer (I wondered if the recent N5171B is compatible). The device is reachable trough Agilent IO connection expert, but the red cross is still visible on BenchVue and I cannot handle it. The ping and RDP connections work well. Does anyone have an idea of what parameter I could change to force the connection on the BenchVue interface ? Many thanks ! |
|
相关推荐
8个回答
|
|
嗨Nicolas,N9020A是支持的频谱分析仪。
但是有几件事情可以让它出现在Connection Expert中而不是BenchVue中。 我首先建议安装最新的IO库(最近发布的17.1)以及BenchVue 2.6,它也刚刚发布。 如果您仍有问题,我们可以进一步深入研究。 但还有一件事值得一试。 有时,NI VISA和Keysight VISA之间的交互可能会导致一些奇怪的问题。 安装最新的IO库可能会解决此问题。 在更新I / O库之前,使用控制面板 - >程序 - >卸载删除共享VISA和IVI组件也可以提供帮助。 还要确保您的N9020A固件是最新的。 您还可以检查它是否可以使用USB,它没有可能出现的局域网问题,只是为了看看你是否可以连接或者是否有其他错误。 最后,有时Appdata文件夹中的信息可能会不正确,因此很难正确设置仪器。 因此,您可以尝试删除整个目录(包括BenchVue本身):( C: Users \ AppData Local Keysight BenchVue Brian 以上来自于谷歌翻译 以下为原文 Hi Nicolas, THe N9020A is a supported Spectrum Analyzer. But there are several things that can make it show up in Connection Expert and not in BenchVue. I would first recommend installing the latest IO Libraries (17.1, recently released) as well as BenchVue 2.6, which was also just released. If you still have trouble, we can drill down further. But one other thing is worth checking. Sometimes interactions between NI VISA and Keysight VISA can cause some odd problems. Installing the newest IO Libraries might fix this issue. Removing the Shared VISA and IVI components using Control Panel->Programs->Uninstall before updating the I/O Libraries can also help. Also make sure your N9020A firmware is up to date. You can also check to see if it works using USB, which doesn't have the LAN issues that can arise, just to see if you can connect at all or if something else is wrong. Finally, sometimes the Appdata folder can get incorrect information in it, making it hard to get the instruments set up correctly. So you can try to delete the entire directory (including BenchVue itself) at: (C:Users\AppDataLocalKeysightBenchVue Brian |
|
|
|
ws2165946 发表于 2018-11-6 14:44 关于N5171B,这是一个信号发生器,而BenchVue还没有支持这类仪器。 但这是我们的路线图,所以请告诉我们您最需要的型号。 我们一直在增加新的。 布赖恩 以上来自于谷歌翻译 以下为原文 Regarding the N5171B, that's a signal generator, and BenchVue does not yet have support for this class of instrument. But it's on our roadmap, so please let us know which models you need the most. We're adding new ones all the time. Brian |
|
|
|
嗨尼古拉斯,这是你可以尝试的另外一件事。
这可能与许多客户有关,他们在BenchVue的LAN仪器上看到红色的X,但可以从Connection Expert中与仪器进行良好的通信。 当BenchVue尝试在LAN上定位仪器时,它会请求LXI识别记录,有时仪器不会响应有用的信息,例如型号。 我们不经常做* IDN吗? 如果我们无法识别它,请确定仪器类型,因为大型LAN上可能有很多仪器。 但你可以强制* IDN? 通过右键单击仪器栏中的仪器,然后单击“刷新”文本来发出。 请试试这个,如果能解决您的问题,请在论坛上告诉我们。 布赖恩 以上来自于谷歌翻译 以下为原文 Hi Nicolas, Here's one other thing you can try. This may be relevant to many customers who see the red X on a LAN instrument in BenchVue but can talk to the instrument fine from Connection Expert. When BenchVue attempts to locate instruments on the LAN, it requests LXI identification records, and sometimes the instruments do not respond with useful information, such as model number. We don't routinely do a *IDN? to determine the instrument type if we can't identify it because there could be a lot of instruments on a large LAN. But you can force a *IDN? to be issued by right clicking on the instrument in the instrument bar and then clicking the "Refresh" text. Please try this and let us all know here on the forum if that solves your problem. Brian |
|
|
|
ws2165946 发表于 2018-11-6 15:18 嗨布赖恩,谢谢你的回复! 我刚刚开始了BenchVue的更新。 奇怪的是它看起来我最近从Keysight网站下载了它,可能在2015-02-19之前...安装2.6后,我甚至看不到MXA或其他仪器在板上。 我删除了AppData文件夹(并附上了我删除之前的所有三个日志)并在屏幕上再次获取。 接下来,我做了刷新命令,但很遗憾无法连接到最新的N9020A(A14.59)。 我们习惯做* IDN? 使用我们的Agilent IO库(64位版本16.3.17914.4)。 我们尝试使用版本17,因为我们收到了带有MXA的CD,但它看起来非常不同,当我们第一次启动它时没有出现相同的连接问题(没有检测到仪器)。 由于在我们这边更新它是一项巨大的工作,我们选择等待一段时间。 我将尝试使用USB线并告诉您结果。 非常感谢和亲切的问候, Keysight BenchVue.log55.7 KBKeysight BenchVue频谱分析仪.log2.6 KB 以上来自于谷歌翻译 以下为原文 Hi Brian, thanks for your replies ! I just have started the BenchVue updates. Strange it looks I recently downloaded it from the Keysight site, probably before 2015-02-19... After installing the 2.6, I do not see even the MXA or another instrument in the board. I deleted the AppData folder (and attached all the three logs I had before deleting them) and get it again on the screen. Next, I did refresh command but it was unfortunately unable to connect to the N9020A which is up to date (A14.59). We are used to do *IDN? with our Agilent IO libraries (64bits version 16.3.17914.4). We tried the version 17 as we received a CD with the MXA but it looks very different and it seemed the same connection problem appeared when we firstly started it (no instrument detected). As it is a huge work to update it on our side, we chose to sill wait a while. I will try with an USB cable and tell you the result. Many thanks and kind regards, 附件
|
|
|
|
bvtwerwer 发表于 2018-11-6 15:28 您好Nicolas,如果您没有看到任何乐器,可能是因为您在BenchVue中启用了“收藏夹”但未在Connection Expert中定义任何乐器。 在这个最新版本中,BenchVue可以只显示您最喜欢的乐器。 如果您的公司局域网中有许多仪器,这可以非常节省时间。 BenchVue右下角的星形图标将打开或关闭收藏夹。 布赖恩 以上来自于谷歌翻译 以下为原文 Hi Nicolas, If you are not seeing any instruments, it may be because you have "Favorites" enabled in BenchVue but have not defined any in Connection Expert. In this latest release, BenchVue can be enabled to show only your favorite instruments. This can be very time-saving if you have a corporate LAN with lots of instruments. The star icon in the lower right corner of BenchVue will turn Favorites on or off. Brian |
|
|
|
ws2165946 发表于 2018-11-6 15:43 嗨比兰,不,我没有激活两边的喜欢,甚至没有在右下角的星形图标。 此外,我再次检查,并没有设法通过网络获得我们的MXA连接。 以上来自于谷歌翻译 以下为原文 Hi Biran, no I have no favorite activated on both sides, and not even a star icon in the lower right corner. Furthermore, I checked again, and did not managed to get a connection on our MXA though the network. |
|
|
|
嗨,Brian,我几乎无法理解刚刚发生的事情;-) MXA现在已被认可....我希望每次将benchvue应用程序安装在另一台机器上时,我都不必将其连接到USB中...... 以上来自于谷歌翻译 以下为原文 Hi Brian, I can barely understand what have just happenned ;-) The MXA is now recognized.... I only hope I won't have to connect it in USB each time the benchvue application is instéalled on another machine... |
|
|
|
bvtwerwer 发表于 2018-11-6 16:03 很高兴MXA现在出现了。 USB修复它,还是你还在使用局域网? 您没有看到“收藏夹”图标的原因是因为在IO库的17.0版本中添加了此功能,并且您表示您现在决定坚持使用16.3。 局域网的另一个问题是在16.3中,您必须手动添加LAN设备。 在ver。 17.x,它们会被自动发现。 如果您添加了MXA并使用了套接字而不是“仪器”设置,则可能导致BenchVue无法查看该仪器。 这是切换到17.x的一个原因。 自动LAN发现可以为您节省大量时间。 但是,正如我之前提到的,如果你的局域网上有很多乐器,那么用收藏夹缩小它也是一件好事。 布赖恩 以上来自于谷歌翻译 以下为原文 Glad the MXA is now showing up. Did USB fix it, or are you still using LAN? The reason you don't see the "Favorites" icon is because this feature was added in version 17.0 of the IO Libraries, and you indicated that you had decided to stick with 16.3 for now. Another thing regarding LAN is that in 16.3, you had to manually add LAN devices. In ver. 17.x, they are automatically discovered. If you added the MXA and used, say, sockets instead of "instrument" settings, that could have accounted for the inability of BenchVue to see the instrument. That is one reason to switch to 17.x. Auto-LAN discovery can save you a lot of time. But, as I mentioned before, if your LAN has a lot of instruments on it, narrowing it down with Favorites is also a good thing. Brian |
|
|
|
只有小组成员才能发言,加入小组>>
1285 浏览 0 评论
2373 浏览 1 评论
2192 浏览 1 评论
2064 浏览 5 评论
2948 浏览 3 评论
1105浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
752浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
923浏览 0评论
1285浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-23 06:05 , Processed in 1.723409 second(s), Total 61, Slave 55 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号