一个是非常暗淡的,另一个是缺少水平线的点(证实它是管),另一个有一些随机的非常低强度的点。 我想购买一个新的显示器,P / N 2090-0217,但在零件网页上,您收到此消息“无法订购,请联系Keysight进行维修服务”。 我意识到这个管子稍微改变了占地面积并且需要一些返工才能使它工作,但它可以做到。 有没有办法订购P / N 2090-0217? 也许签署“我承担所有责任”豁免? 谢谢,VictorEdited:vjsilva1于2014年8月23日上午7:41发布 以上来自于谷歌翻译 以下为原文 I have a few 3458As that need a new display. One is very dim, another has a missing horizontal line of dots (verified it's the tube) and another has some random very low intensity dots. I would like to purchase a new display, P/N 2090-0217 but in the parts webpage you get this message "Not orderable, Contact Keysight for repair service". I realize this tube has changed the footprint slightly and requires some rework to make it work, but it's possible to do. Is there any way to order P/N 2090-0217? Maybe signing a "I assume all reponsibility" waiver? Thanks, Victor Edited by: vjsilva1 on Aug 23, 2014 7:41 AM 附件
|
|
相关推荐
2个回答
|
|
我希望安捷伦将Keysight分拆出来,这个荒谬的政策“不可订购,联系Keysight维修服务”将被淘汰?
以上来自于谷歌翻译 以下为原文 I'm hoping with the spin off of Keysight from Agilent that this ridiculous policy "Not orderable, Contact Keysight for repair service" will be eliminated?? |
|
|
|
嗨,我想我会把它扔到那里寻找可能的解决方案。
如果您正在使用Keysight作为3458A的校准提供商,您可以考虑购买他们的服务协议以及校准服务。 除校准成本外,2年期协议约为350美元。 一旦他们满意3458A正常运行,服务协议应涵盖更换昏暗的显示器。 我曾经要求他们解决3458A的某些服务说明,例如Cal Ram电池问题,因为他们在Loveland标准实验室拥有DMM,他们一直在做,他们已经取代了本服务协议所涵盖的昏暗显示器。 好。 如果您还没有提到服务说明15D,您可能需要考虑它,因为这可能会在某些时候出现痛苦的失败,特别是如果您的3458A更老了。 祝你好运,兰迪 以上来自于谷歌翻译 以下为原文 Hi, I thought I would throw this out there for a possible solution for you. If by chance you are using Keysight as a calibration provider for your 3458A's you might consider purchasing their service agreement along with the calibration service. A 2 year agreement is about $350.00 on top of the calibration cost. Once they are satisfied the 3458A is functioning properly the service agreement should cover the replacement of the dim displays. I have asked them in the past to address certain service notes for the 3458A like the Cal Ram Battery issue when they have the DMM at the Loveland standards lab which they've always done, and they have replaced dim displays covered under this service agreement as well. If you haven't already addressed service note 15D you might want to consider it as this is potentially a painful failure at some point especially if your 3458A's are older. Good luck with it, Randy |
|
|
|
只有小组成员才能发言,加入小组>>
1285 浏览 0 评论
2373 浏览 1 评论
2192 浏览 1 评论
2064 浏览 5 评论
2948 浏览 3 评论
1108浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
752浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
924浏览 0评论
1287浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-23 14:16 , Processed in 1.606113 second(s), Total 81, Slave 64 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号