嗨 - 很可能,你看到的是两个MXA之间绝对幅度精度的差异。
在不知道被测信号的标称幅度和频率的情况下,很难说您所看到的是否符合规范。
例如,如果您在50 MHz处测量-20 dBm信号,则N9020A在20至30摄氏度范围内的绝对振幅规格为+/- 0.33 dB。
这意味着测量相同50 MHz信号的两个分析仪理论上可以显示高达0.66 dB的差异,并且都在规格范围内。
但要在这种极端情况下使用两台分析仪的可能性很小。
但是,如果此测量是在5 GHz进行的,则绝对幅度精度规格为+/- 0.33 dB + 1.5 dB,总计+/- 1.83 dB。
测量同一信号的两个MXA可能会显示+/- 3.66 dB的差异。
要在此频率下看到0.5 dB和0.7 dB之间的差异是完全合理的。
问候 -
以上来自于谷歌翻译
以下为原文
Hi -
Very likely, what you are seeing is the difference in absolute amplitude accuracy between the two MXAs. It is difficult to say whether what you are seeing is within specification, or not, without knowing the nominal amplitude and frequency of the signals being measured. For example, if you are measuring a -20 dBm signal at 50 MHz, the absolute amplitude specification for the N9020A over a 20 to 30 degC range is +/- 0.33 dB. This means that two analyzers measuring this same 50 MHz signal could theoretically show a difference of as much as 0.66 dB and both be within specification. But to have two analyzers at such extremes is highly unlikely.
However, if this measurement were being made at 5 GHz, the absolute amplitude accuracy spec would be +/-0.33 dB + 1.5 dB for a total of +/-1.83 dB. Two MXAs measuring this same signal could show a difference of as much as +/-3.66 dB. To see a difference of between 0.5 dB and 0.7 dB at this frequency is entirely reasonable.
Regards -
嗨 - 很可能,你看到的是两个MXA之间绝对幅度精度的差异。
在不知道被测信号的标称幅度和频率的情况下,很难说您所看到的是否符合规范。
例如,如果您在50 MHz处测量-20 dBm信号,则N9020A在20至30摄氏度范围内的绝对振幅规格为+/- 0.33 dB。
这意味着测量相同50 MHz信号的两个分析仪理论上可以显示高达0.66 dB的差异,并且都在规格范围内。
但要在这种极端情况下使用两台分析仪的可能性很小。
但是,如果此测量是在5 GHz进行的,则绝对幅度精度规格为+/- 0.33 dB + 1.5 dB,总计+/- 1.83 dB。
测量同一信号的两个MXA可能会显示+/- 3.66 dB的差异。
要在此频率下看到0.5 dB和0.7 dB之间的差异是完全合理的。
问候 -
以上来自于谷歌翻译
以下为原文
Hi -
Very likely, what you are seeing is the difference in absolute amplitude accuracy between the two MXAs. It is difficult to say whether what you are seeing is within specification, or not, without knowing the nominal amplitude and frequency of the signals being measured. For example, if you are measuring a -20 dBm signal at 50 MHz, the absolute amplitude specification for the N9020A over a 20 to 30 degC range is +/- 0.33 dB. This means that two analyzers measuring this same 50 MHz signal could theoretically show a difference of as much as 0.66 dB and both be within specification. But to have two analyzers at such extremes is highly unlikely.
However, if this measurement were being made at 5 GHz, the absolute amplitude accuracy spec would be +/-0.33 dB + 1.5 dB for a total of +/-1.83 dB. Two MXAs measuring this same signal could show a difference of as much as +/-3.66 dB. To see a difference of between 0.5 dB and 0.7 dB at this frequency is entirely reasonable.
Regards -
举报