[问答]

如何查看我写的SCPI命令是否正确

大家好,我有两个问题想问这里:1。如何查看我写的SCPI命令是否正确,我在哪里可以运行它?
2.我想使用PULSE进行一些测量,但我刚刚发现了我正在使用的SCPI命令:SOURce [1]:FUNCtion:SHAPe“不支持Agilent B2901A仪器我从安捷伦获得命令
B2900系列SCPI命令参考页325.我只是没有得到它,因为仪器具有该功能并且我已成功手动测试为什么没有SCPI命令?谢谢

以上来自于谷歌翻译


     以下为原文

  Hello all,

I have two questions want to ask here:

1. How can I check the SCPI commands I wrote is right or not, where can I run it?

2. I want to use PULSE to do some measurement,but I just found out this SCPI command I am using

    :SOURce[1]:FUNCtion:SHAPe" is not supported by the instrument which is Agilent B2901A  

    I get the command from the Agilent B2900 Series SCPI command reference page 325.

    I just do not get it since the instrument has that functionality and I been manually tested successfully why there is no SCPI command for that?

Thanks
附件

回帖(1)

杨梅

2018-12-11 16:10:20
与熟悉该产品的工程师交谈,他们不得不说:B2901A / 02A / 11A / 12A支持两个遥控命令集。
一个是支持所有B2901A / 02A / 11A / 12A功能的默认命令集。
另一种是传统的命令集,其设计用于使用客户创建的现有程序来控制现有仪器,例如Keithley 2400.仪器可以配置为使用传统的命令集。
首先,请按照以下链接中描述的步骤检查仪器上的遥控器命令设置。
http://www.home.agilent.com/agilent/editorial.jspx?ckey=2137580&id=2137580&nid=-33504.0.00&lc=eng&cc=US然后,它被配置为使用
常规命令集,请通过以下链接中描述的步骤选择仪器上的默认命令集。
http://www.home.agilent.com/agilent/editorial.jspx?cc=US&lc=eng&ckey=2137603&nid=-33504.0.00&id=2137603之后,如果您仍然遇到问题
仪器无法接收“:SOURce [1]:FUNCtion:SHAPe”命令,请告知我们更详细的信息。
下面的编程指南可能有助于客户除了SCPI命令参考之外还使用PULSE。
它显示了编程示例。
http://cp.literature.agilent.com/litweb/pdf/B2910-90020.pdf

以上来自于谷歌翻译


     以下为原文

  spoke with an engineer who is knowledgeable on this product and this is what they had to say:

The B2901A/02A/11A/12A supports two remote control command sets.
One is the default command set which supports all B2901A/02A/11A/12A functions.
The other is the conventional command set designed for using the existing programs 
which the customer created to control existing instruments such as Keithley 2400.

The instrument may be configured to use the conventional command set.
So first of all, please check the remote control command setting on the instrument
by the procedure described in the link below.
http://www.home.agilent.com/agilent/editorial.jspx?ckey=2137580&id=2137580&nid=-33504.0.00&lc=eng&cc=US

And then, it is is configured to use the conventional command set,
please select the default command set on the instrument by te procedure described in the link below.
http://www.home.agilent.com/agilent/editorial.jspx?cc=US&lc=eng&ckey=2137603&nid=-33504.0.00&id=2137603

After that, if you still face the issue which the instrument can't receive ":SOURce[1]:FUNCtion:SHAPe" command,
please let us know the more detailed information.
Programming Guide below may be helpful for the customer to make use of PULSE 
in addition to SCPI command reference. It shows programming examples.
http://cp.literature.agilent.com/litweb/pdf/B2910-90020.pdf