[问答]

校准通过SCPI命令响应

嗨,我正在使用网络分析仪:E5071C我在查询SCPI命令(使用labview)时将功率电平归零,同时尝试使用cal方法作为响应(Thru)手动相同的事情将设置如下:Cal >> Calibrate>
>响应(直通)>> SelectPorts(2,1)>>通过>>完成我试图使用keysight网站提供的帮助找到相应的命令:http://ena.support.keysight.com/e5071c/manuals/
webhelp / eng /网络anaylzer上的软键:等价SCPI选择端口:SENS1:CORR:COLL:THRU 2,1通过:SENS1:CORR:COLL:THRU 2,1; * OPC?
完成:SENS1:CORR:COLL:SAVE“我将配置所需的配置如下:CALC1:PAR1:DEF S21:SENS1:SWE:TYPE POW:SOUR1:POW:STAR -45:SOUR1:POW:STOP 5:SENS1:
FREQ 2620000000:SENS1:SWE:POIN 201:SENS1:BAND 1E3然后尝试运行命令:SENS1:CORR:COLL:THRU 2,1出错:*需要额外的标准。*请帮我找到正确的方法
使用LabVIEW或SCPI命令集执行Cal >> Resp(thru)方法。提前谢谢!关于Geeta Kannammanavar

以上来自于谷歌翻译


     以下为原文

  Hi,

I am using Network analyzer: E5071C
I have query on SCPI command(using LabVIEW) to zeroing the power level while trying to use the cal method as response(Thru)

Manually same thing will be set as below:

Cal>>Calibrate>>Response(Thru)>>SelectPorts(2,1)>>Thru>>Done

I tried to find the equivalent command using the help available in keysight website:
http://ena.support.keysight.com/e5071c/manuals/webhelp/eng/


Soft Keys on Network anaylzer:     Equicvalent SCPI
Select Ports      :SENS1:CORR:COLL:THRU 2,1
Thru                     :SENS1:CORR:COLL:THRU 2,1;*OPC?
Done                     :SENS1:CORR:COLL:SAVE"


I will configure the required config as below  

:CALC1:PAR1:DEF S21
:SENS1:SWE:TYPE POW
:SOUR1:POW:STAR -45
:SOUR1:POW:STOP 5
:SENS1:FREQ 2620000000
:SENS1:SWE:POIN 201
:SENS1:BAND 1E3

Then try to run the command  
:SENS1:CORR:COLL:THRU 2,1
Getting an error : *Additional standard needed.*

Please help me to find the right way to execute the Cal>>Resp(thru) method using LabVIEW or SCPI command set.

Thank You in advance!.

Regards
Geeta Kannammanavar  

回帖(2)

贾永世

2018-11-6 10:49:53
亲爱的Geeta,错误“需要附加标准”表明,在完成计算校准系数所需的所有数据测量之前,已经发送了打开校准功能的GPIB命令。
如果你想通过校准执行响应,请尝试定义如下,SENS1:CORR:COLL:METH:THRU 1,2 SENS1:CORR:COLL:THRU 1,2 * OPC?
SENS1:CORR:COLL:SAVE

以上来自于谷歌翻译


     以下为原文

  Dear Geeta,

The error "Additional standard needed" shows that the GPIB command that turns ON the calibration function has been sent before all the data measurements required to calculate the calibration factor have been completed.

if you want to perform response through calibration, please try to define as follows,

SENS1:CORR:COLL:METH:THRU 1,2
SENS1:CORR:COLL:THRU 1,2
*OPC?
SENS1:CORR:COLL:SAVE
举报

刘勉励

2018-11-6 11:02:36
引用: 大洼球王 发表于 2018-11-6 20:28
亲爱的Geeta,错误“需要附加标准”表明,在完成计算校准系数所需的所有数据测量之前,已经发送了打开校准功能的GPIB命令。
如果你想通过校准执行响应,请尝试定义如下,SENS1:CORR:COLL:METH:THRU 1,2 SENS1:CORR:COLL:THRU 1,2 * OPC?
SENS1:CORR:COLL:SAVE

谢谢Takeshio!

以上来自于谷歌翻译


     以下为原文

  Thank You Takeshio !
举报

更多回帖

发帖
×
20
完善资料,
赚取积分