您好!有人可以帮我理解校准和调整的意义。在波形发生器33210A的手册中,它说:“对于大多数应用来说,1年间隔是足够的。只有在调整时,精度规格仍然有效。
定期校准间隔。精度规格在1年校准间隔后无效。安捷伦建议不要将校准间隔延长超过2年。规格仅在上次调整规定的时间内有效。无论您选择何种校准间隔, 安捷伦建议始终在校准间隔内进行完全重新调整,这对于确保仪器在下一个校准间隔内保持在规格范围内是必要的。重新调整的标准提供了最佳的长期稳定性。 性能验证测试并不意味着仪器将保留在这些中 除非进行调整,否则将进行调整。“因此,校准应在1年内进行,调整情况如何?我的问题是,如果您在一年或两年内校准设备,是否需要内部重新调整?或内部重新调整是 在内部进行校准的方法?感谢您的帮助,问候,André 以上来自于谷歌翻译 以下为原文 Hello! Can anybody please help me understand the calibration and adjustment meaning. In the manual of the 33210A, Waveform generator, it said: "A 1-year interval is adequate for most applications. Accuracy specifications will remain valid only if adjustment is made at regular calibration intervals. Accuracy specifications are not valid beyond the 1-year calibration interval. Agilent does not recommend extending calibration intervals beyond 2 years for any application. Specifications are valid only within the period stated from the last adjustment. Whatever calibration interval you select, Agilent recommends that complete re-adjustment always be performed at the calibration interval. This is necessary to ensure that the instrument will remain within specification for the next calibration interval. This criteria for re-adjustment provides the best long-term stability. Performance data measured during Performance Verification Tests does not mean that the instrument will remain within these limits unless the adjustments are performed." So, calibration should be done in 1-year and what about adjustments? My question is, if you calibrate the equipament during one year or two, does it needs internal readjustments? or the internal readjustments are a way to do the calibration internally? Thank you for the help, Regards, André |
|
相关推荐
1个回答
|
|
嗨,安德烈!安捷伦建议您在33210A上进行校准时始终执行完全重新调整。
您无需打开仪器即可进行调整 - 所有操作均以电子方式完成。性能规格为自上次调整设备后的1年。 您可以选择使用您的应用所需的校准间隔,但安捷伦建议不要超过2年。 我希望有所帮助! 以上来自于谷歌翻译 以下为原文 Hi, Andre! Agilent recommends that you always perform a full re-adjustment whenever a calibration is performed on the 33210A. You do not have to open the instrument to perform adjustments -- it is all done electronically. The performance specifications are for a period of 1 year from the last time the unit was adjusted. You can choose to use whatever calibration interval you need for your application, but Agilent does not recommend exceeding 2 years. I hope that helps! |
|
|
|
只有小组成员才能发言,加入小组>>
1287 浏览 0 评论
2375 浏览 1 评论
2194 浏览 1 评论
2064 浏览 5 评论
2951 浏览 3 评论
1112浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
753浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
927浏览 0评论
1287浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-24 10:33 , Processed in 1.501026 second(s), Total 66, Slave 60 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号