假设光圈时间实际上不等于测量时间,我是否正确?
我问,因为即使光圈为500ns,我也无法测量10us脉冲的峰值。 DSDC模式工作正常,但我失去了自动量程,这是DCV模式的一个很好的功能。 我已经用尽了手册中有关DCV模式数字化的所有选项,但似乎无法找到一种方法来测量任何小于25us长脉冲的峰值。 即便如此,我还是要将测量延迟到脉冲中以获得正确的读数。 几乎看起来好像我的信号有一个~40kHz的低通。 如果有人有任何想法,我很乐意听到他们。 谢谢! 以上来自于谷歌翻译 以下为原文 Am I correct in assuming that aperture time does not actually equal measurement time? I'm asking because I cannot measure the peak value of a 10us pulse even with with an aperture of 500ns. DSDC mode works fine, but I lose the auto ranging which is a nice feature of DCV mode. I've exhausted all of the options in the manual regarding digitizing with DCV mode, but can't seem to find a way to measure a peak value on anythign less than a 25us long pulse. And even then I have to delay the measurement 20us into the pulse to get a correct reading. It almost seems as if there is a ~40kHz low pass on my signal. If anyone has any ideas I would love to hear them. Thanks! |
|
相关推荐
1个回答
|
|
您好,您目前如何设置3458A,也许您可以提供命令列表?
你是正确的,光圈不等于每次测量的总时间。 前者有很多东西可以发挥作用:自动归零,自动量程,配置时间,总线时间(如果编程)等等......光圈是实际信号的测量时间,但这只是其中的一部分。 处理。 现在DCV模式下的最大读取速度为100Ksamples / sec,设置需要恰到好处才能达到该速度,并且您的信号为100KHz,即每个周期只有1个样本。 你达到巅峰的可能性不是很大。 通常我建议数字化信号然后只是打开数学并查看最大最小值,或者自己对数据进行后处理,但如果你的信号是100KHz,那就不行了。 为了满足奈奎斯特,你的信号必须是50KHz或更低,而这正在推动它。 另外注意,你是正确的DSDC模式在捕获峰值方面比DCV更好。 你说你喜欢DCV,因为自动量程,但是在DCV中,如果你的自动量程,你将无法快速采样以捕获你想要的脉冲。 您必须使用手动范围。 如果你提供更多信息,我会很乐意尽我所能,但我不承诺你能在DCV模式下做你想做的事。 以上来自于谷歌翻译 以下为原文 Hello, How do you currently have the 3458A setup, maybe you could provide a command list? You are correct that aperture does not equal total time per measurement. There is a lot of things that play into that ex: autozero, autorange, config time, bus time (if programming), etc.... The aperture is how long the actual signal is measured, but that is just one part of the process. Now the max readspeed in DCV mode is 100Ksamples/sec, and the settings need to be just right to achieve that speed, and your signal is 100KHz, that's only 1 sample per cycle. The chances of you getting the peak is not very good. Normally I suggest digitizing a signal then just turning on math and looking at the max min, or post process the data yourself, but if your signal is 100KHz, that's not going to work. In order to fulfill nyquist your signal would have to be 50KHz or less, and that's pushing it. One other note, you are correct DSDC mode is better at capturing peaks than DCV. You say you like DCV because of autoranging but, in DCV, if your autoranging there is no way your gonna sample quickly enough to capture the pulse you want. You must use a manual range. I'll be happy to help best I can if you provide some more information, but I'm not promising you'll be able to do what you want in DCV mode. |
|
|
|
只有小组成员才能发言,加入小组>>
1294 浏览 0 评论
2377 浏览 1 评论
2199 浏览 1 评论
2068 浏览 5 评论
2955 浏览 3 评论
1131浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
759浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
939浏览 0评论
1296浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-27 01:48 , Processed in 1.180498 second(s), Total 47, Slave 40 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号