在与U1253A有不同的问题(并且没有得到回复)之后,我注意到了仪表的一些问题。
第一个是当我在测量带有空选项的电容器的情况下,即使范围本身确实根据上限值改变,范围图标也不会改变。 这是次要的,并没有真正影响测量,所以我可以让这一个去。 然而,仪表有一个更严重的问题:我用它来测量PS的输出纹波顺序你知道这个问题的原因是什么以及如何修复它? 当所有改变后的测量之间的这种误差是背光设置似乎是一个主要问题。 以上来自于谷歌翻译 以下为原文 After having a different issue with the U1253A (and getting no reply), I've noticed a few more problems with the meter. The first is the fact when I'm measuring capacitors with the null option on, the range icon doesn't change even though the range itself does change according to the cap value. This is minor and not really affecting the measurement so I can let this one go. However, there's a much more serious issue with the meter: I'm using it to measure output ripple of PS's in the order of <1mV. I'm using a fully charged battery, and the meter is acting up. For example, with the last PS I've measured, the meter reads about 0.6mVAC when the back light is at its max or medium position, and about 0.4-0.5mV (I can't remember exactly what it was, as the measurment was taken a couple of weeks ago) when the back light is at the low setting. This is a very high error, about 30% between the 2 measurements! I don't have a good DMM at hand to compare to so I can't even tell which one is correct. I've used a keithley 2010 at one of the labs at work and it reads about 0.4mV, but this is using different probes, and at different time and place so the two measurements cannot be directly compared. Especially since my U1253A wasn't calibrated in over 2 years. Do you know what is the cause of this problem and how it can be fixed? Such an error between measurements when all that is changed is the back light setting seems like a major issue. |
|
相关推荐
6个回答
|
|
这么久没有回复的事实意味着这个问题没有任何解决方案/答案?
以上来自于谷歌翻译 以下为原文 The fact there's no reply for so long means there won't be any solution/answer to the issue? |
|
|
|
在U1253A上进行了几次测试ACV的测试,没有发现问题(NTF)。
在测量0.6mV的输出数值时,我也打开了半小时的背光。 根本没有问题,ACV保持在规范范围内。 我想向您保证,背光打开或关闭时,它与测量ACV无关。 通过对另一个单元的进一步调查,我仍然无法模拟影响ACV的背光功率的情况。 在重申中,结果是NTF,你提到的问题没有被观察到。 我建议您联系Agilent Online技术支持以获取更多信息。 以上来自于谷歌翻译 以下为原文 Several tests were ran for measuring ACV on the U1253A, there is no trouble found (NTF). I have also turned on the backlight for half an hour while measuring an output figure of 0.6mV. There is no issue at all, ACV stays within the specification. I like to assure you that with the backlight turned on or off, it has no relation to measuring ACV. By further investigation with another unit, I still cannot simulate the case of back light power affecting the ACV. In the reiteration, the result is NTF, the issue you mentioned is not observed. I would like to suggest that you contact Agilent Online technical support for more information. |
|
|
|
就这么清楚,我的意思是亮度,而不是背光,因为U1253A(OLED)上没有背光 - 对于滥用这个词感到抱歉。你能告诉我你是如何测量它的吗?
我实际上没有输入0.6mV的交流电,我测量了稳压电源的输出纹波,这可能低于此值。 你能尝试相同的测试吗? 我得到的所有PS都是相同的,甚至我的旧凳子PS(Horizon dhr 40-1)。 以上来自于谷歌翻译 以下为原文 Just so were clear, I meant the brightness and not the back light as there' no back light on the U1253A (OLED) - sorry for the misuse of the word. Can you please let me know how you've measured it? I didn't actually input a 0.6mV AC, I've measured the output ripple of a regulated power supply which is probably lower than that. Can you try the same test? I get the same for all the PS I measure even my old bench PS (Horizon dhr 40-1). |
|
|
|
您好,我已经将仪表发送到当地的安捷伦分公司,描述了两个问题 - 电容测量和显示对测量的影响(两者的视频 - 如果您希望我可以将它们上传到YouTube并发布链接
在论坛中,如果附加的链接不起作用)。 根据它们,仪表没有任何问题,这个问题可能会广泛传播到所有这些单位。 这不是你在帖子中描述的内容。 你可以重复相同的测试并验证它仅限于我的单位/广泛的问题吗?以下是视频:http://dl.dropbox.com/u/37472979/Voltage.3gphttp://dl.dropbox.com /u/37472979/Capacitors.3gp 以上来自于谷歌翻译 以下为原文 Hello, I've sent the meter over to the local Agilent branch describing both issues - the capacitance measurement and the display affect on the measurement (with a video of both - if you'd like I can upload them to YouTube and post a link here in the forum in case the attached links won't work). According to them there is nothing wrong with the meter and the issue is probably wide spread to all of these units. Which isn't what you've described in your post. Can you please repeat the same tests and verify its only limited to my unit/a widespread issue? Here are the videos: http://dl.dropbox.com/u/37472979/Voltage.3gp http://dl.dropbox.com/u/37472979/Capacitors.3gp |
|
|
|
有趣的视频。有任何状态更新,这个仪表是否被更换?
以上来自于谷歌翻译 以下为原文 Interesting videos. Is there any status update, did this meter get replaced ? |
|
|
|
抱歉没有更新。
安捷伦告诉我,我的仪表没有问题,这个问题在这个系列的所有型号(U125x)中都很普遍。 根据EEVblog的一些人(我也发布过),他们的U1252A没有电容问题(也可能是准确性问题),所以我猜安捷伦的官方回应并非100%正确, 无论如何,他们已经把我的米送回了。我不能说我没有多少失望,我没想到这么昂贵的仪表会出现这样的问题。 我可以说的是,当地安捷伦分公司的员工非常友好并且确实试图帮助我。 我希望他们至少能更换仪表,以确保这个问题确实普遍存在(他们在当地分公司没有任何这些HHDMM所以他们已向国际办公室询问了这个问题),但它没有 发生。 以上来自于谷歌翻译 以下为原文 Sorry for not updating. Agilent has informed me there's no problem with my meter, and this problem is widespread in all of the models of this series (U125x). According to some people at EEVblog (where I've also posted), their U1252A doesn't have the capacitance issue (and probably the accuracy issue as well), so I guess the official response from Agilent isn't 100% correct, but they've sent me my meter back anyway. I can't say I'm not somewhat disappointed, I've wasn't expecting such issues from such an expensive meter. All I can say is that the people at the local Agilent branch were very nice and did try to help me. I've hoped they would at least replace the meter to be sure this issue is indeed widespread (they don't have any of these HHDMM at the local branch so they've asked the international office about it), but it didn't happen. |
|
|
|
只有小组成员才能发言,加入小组>>
1230 浏览 0 评论
2351 浏览 1 评论
2160 浏览 1 评论
2026 浏览 5 评论
2908 浏览 3 评论
974浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
707浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
808浏览 0评论
1230浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-11-26 05:21 , Processed in 1.531099 second(s), Total 88, Slave 71 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号