您好,Anybody有一段关于如何通过E5061A校准进行传输的代码,其中有四条迹线被激活?
这让我疯狂,手册没什么帮助。 此代码不起作用。 分析仪发出哔哔声,但屏幕上只显示两条迹线和两个耗散器。 然而,如果我手动按下分析仪上的传输按钮,它会校准所有四条迹线。任何想法? 使用示例10 OUTPUT 717;“:SENS1:CORR:COLL:THRU 1,2”20 OUTPUT 717;“* OPC?”30 ENTER 717;等效键[Cal] - 校准 - 响应(直通) - 直通[Cal] - Calibrate - n-Port Cal - Transmission - Port mn Thru 以上来自于谷歌翻译 以下为原文 Hello, Anybody has a piece of the code on how to do a transmission thru calibration on the E5061A with four traces activated? This is driving me nuts and the manual is not much help. This code does not work. The analyzers beeps but only two traces show on the screen and two dissipater. however If i manually press on the transmission thru button on the analyzer it calibrates all four traces. Any ideas? Example of use 10 OUTPUT 717;":SENS1:CORR:COLL:THRU 1,2" 20 OUTPUT 717;"*OPC?" 30 ENTER 717;A Equivalent key [Cal] - Calibrate - Response (Thru) - Thru [Cal] - Calibrate - n-Port Cal - Transmission - Port m-n Thru |
|
相关推荐
2个回答
|
|
也许你可以执行“SYST:ERR?”
查询SCPI发送后是否存在错误。 我怀疑发送了错误的SCPI。 以上来自于谷歌翻译 以下为原文 Perhaps you can perform "SYST:ERR?" to query whether error exist on instrument after those SCPI were sent. i am suspects that wrong SCPI were sent. |
|
|
|
你必须首先定义校准方法,即。
“SENS1:CORR:COLL:METH:THRU 2,1”“SENS1:CORR:COLL:THRU 2,1”“SENS1:CORR:COLL:SAVE” 以上来自于谷歌翻译 以下为原文 You have to define the method of calibration first ie. "SENS1:CORR:COLL:METH:THRU 2,1" "SENS1:CORR:COLL:THRU 2,1" "SENS1:CORR:COLL:SAVE" |
|
|
|
只有小组成员才能发言,加入小组>>
1291 浏览 0 评论
2375 浏览 1 评论
2194 浏览 1 评论
2065 浏览 5 评论
2952 浏览 3 评论
1119浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
756浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
931浏览 0评论
1293浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-25 21:50 , Processed in 1.592421 second(s), Total 79, Slave 63 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号