嗨,我们正在开发一个labview应用程序(使用LV2011)来远程控制E5061B。
我们使用可用的ENA系列仪器驱动程序,大部分功能都应该如此。 当E5061B检测到内部错误时,会出现通信问题。 如果我开始描述我们的功能目标,也许它会为其他人提供更多信息。 我们想测试对象是否存在E5061B在R / T端口检测到输入过载的风险。 然后当仪器检测到这样的错误时,我们想要重置错误并用校正的设置开始新的扫描(低功率,输入Att。)。 我发现,当检测到错误时,仪器被阻止进行远程控制。 我想重置仪器,它在发送电子设备时不会做出正常的反应。 * RST字符串命令。 如果有人能给我一些建议,我会很高兴吗? Oddgeir 以上来自于谷歌翻译 以下为原文 Hi, We are developing a labview application (using LV2011) for remote controlling of the E5061B. We use the available ENA series instrument drivers and mostly it functions as it should. The communication problem occours when E5061B detect an internal error. Maybe it more informative for others if I start to describe our aim of funtionality. We want to to test on object that we can have a risk that the E5061B detect input overload at R/T-ports. Then when the instrument detect such a error, we want to reset the error(s) and start a new sweep with corrected setup (lower power, input Att.). It semms to me that the instrument is blocked for remote controlling when the error is detected. I want to reset the instrument, it do not react normaly when sending the e.g. *RST string command. I will be happy if anyone can give me some advise to do? Oddgeir |
|
相关推荐
1个回答
|
|
Hello Oddgeir如果您有E5061B选项3L5,它具有独特的过载保护功能。
当E5061B 3L5检测到过载时,它会改变仪器设置并保护内部电路。 http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/setting_control_functions/overload_detection_and_power_trip_function.htm您需要使用SCPI命令清除过载保护状态“:SERVICE:POWER:OVERload:PROTect:CLEar”http: //ena.tm.agilent.com/e5061b/manuals/webhelp/eng/programming/command_reference/service/scpi_service_power_overload_protect_clear.htm但我担心,如果你有这种情况,Labview应用驱动程序可能会阻止该工具。 如果您无法向E5061B发送任何SCPI命令,请联系NI支持团队以了解如何重置被阻止的通信。 以上来自于谷歌翻译 以下为原文 Hello Oddgeir If you have E5061B option 3L5, this has unique overload protection functions. When E5061B 3L5 detects overload, it changes instrument setting and protect internal circuit. http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/setting_control_functions/overload_detection_and_power_trip_function.htm You need to clear overload protection status with a SCPI command ":SERVice:POWer:OVERload:PROTect:CLEar" http://ena.tm.agilent.com/e5061b/manuals/webhelp/eng/programming/command_reference/service/scpi_service_power_overload_protect_clear.htm But I am afraid that the instrument may be blocked by Labview appliction of driver in the case you have. If you can't send any SCPI command to E5061B, please contact NI support team to know how to reset the blocked communication. |
|
|
|
只有小组成员才能发言,加入小组>>
1067 浏览 0 评论
2305 浏览 1 评论
2092 浏览 1 评论
1966 浏览 5 评论
2837 浏览 3 评论
833浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
614浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
711浏览 0评论
1395浏览 0评论
1067浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-9-29 04:55 , Processed in 1.395257 second(s), Total 77, Slave 61 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号