您好论坛,我每隔5-15分钟就会经常遇到一个Align Now Required状态消息。
由于该分析仪支持的测试装置,我使用手动Align Now,All功能,而不是自动对齐。 因此,我会不断轮询此状态并在需要时执行对齐。 这显着增加了该特定自动测试仪的总测试时间的随机时间量。 这种分析相对较新(1年),该设备位于清洁,温度稳定的实验室环境中,并且处理得当。 有关为何经常需要对齐的建议或见解? 谢谢,戴夫 以上来自于谷歌翻译 以下为原文 Hello Forum, I'm encountering an Align Now Required status message very frequently, anywhere from every 5-15 minutes. Due to the test set this analyzer supports, I am using the manual Align Now, All function, not the auto align. Therefore, I'm constantly polling this status and performing an alignment when required. This significantly adds a random amount of time to the total test time for this specific automated tester. This analysis is relatively new (1 year) and the unit resides in a clean, temperature stable lab environment and has be properly handled. Any suggestions or insight as to why an alignment is required so frequently? Thanks, Dave |
|
相关推荐
3个回答
|
|
嗨Dave - PXA将“要求”每15分钟对齐一次,这是正常的。
许多对齐中的一个(对齐之间的最短时间对齐)大约每15分钟“到期”。 我每隔15分钟提一次“约”,因为时间和温度方程都会触发对齐频率。 情况是,根据安装的选项,PXA可能有超过50种可能的对齐方式,并且这些对齐在不同时间到期。 因此,有时您会看到仪器每15分钟需要对齐一次,有时您会看到另一个对齐(可能是2-3小时的时间表),每15分钟对齐一次。 问候 - 以上来自于谷歌翻译 以下为原文 Hi Dave - The PXA will ‘ask” to be aligned about every 15 minutes, and this is normal. One of the many alignments (the alignment with the shortest time period between aligns) “expires” about every 15 minutes. I mention “about” every 15 minutes because there is both a time and temperature equation that triggers the alignment frequency. The situation is that the PXA may have over 50 possible alignments depending on options installed, and these alignments expire at different times. Therefore sometimes you see the instrument needing alignment every 15 minutes and sometimes you see another alignment (that might be on a 2-3 hour schedule) expiring between the alignments that occur every 15 minutes. Regards - |
|
|
|
醉狼工作室 发表于 2018-12-11 16:24 tabbott,感谢您的快速回复。 还有一个问题。 如果仅在5-10分钟后需要对齐,可能出现的症状有哪些? PXA可能存在问题吗? 谢谢,戴夫 以上来自于谷歌翻译 以下为原文 tabbott, Thanks for the prompt response. One additional question. What are the possible symptoms if an alignment is required after only 5-10 minutes? Is it likely that PXA has an issue? Thanks, Dave |
|
|
|
嗨Dave - 如果PXA出现对齐问题,由于某种原因对齐失败,您将在屏幕左下方的状态栏上看到一条消息,如“IF对齐失败”。 另一个要查看的地方是按系统,显示,错误的错误日志。 如果对齐失败,或者仪器有其他错误,例如参考解锁或SCPI错误,您将在此处看到这些错误的列表。 此外,通常如果自动校准失败,仪器将尝试每1分钟重新运行校准! 您还将在屏幕上看到“立即对齐所有必需”消息,此消息不会消失,因为无法完成对齐。 我怀疑,因为你没有提到Align Now All Required消息没有消失,或者你已经看到任何对齐失败注释出现在屏幕上; 没有PXA硬件问题(您没有损坏的单元)。 我们测量仪器的内部温度,这些温度变化包含在确定何时需要对准的时间和温度“规则”中。 我提到我们监测内部仪器温度,因为我预计环境室温的微小变化不会对仪器内部温度产生1:1的影响。 如果您的设施在夜间关闭空调或加热器,这可能会导致一些温度波动,这些波动会在一天中的某些时段以更高的频率触发自动对齐。 问候 - 以上来自于谷歌翻译 以下为原文 Hi Dave - If the PXA has an alignment issue where the alignment fails for some reason, you will see a message appear on the lower left status bar on screen such as ”IF Alignment failure”. Another place to look is Error log by pressing System, Show, Errors. If an alignment has failed, or the instrument has some other error such as Reference Unlock, or a SCPI error, you will see a list of these errors here. Also, usually if an auto alignment does fail, the instrument will try and re run the alignment every 1 minute! You will also see the Align Now All Required message on screen, and this message will not go away since an alignment cannot be completed. I suspect since you have not mentioned that the Align Now All Required message does not go away, or that you have seen any alignment failure annotation appear on screen; that there are no PXA hardware issues (you do not have a broken unit). We measure the internal temperature of the instrument and these temperature changes are included in the time and temperature “rules” that determine when an alignment is needed. I mention that we monitor in internal instrument temperature since I expect minor changes in ambient room temperature would not have a 1:1 effect on internal instrument temperature. If your facility turns off the air conditioner or heater at night, this could account for some temperature fluctuations that will trigger auto aligns to occur with greater frequency during certain parts of the day. Regards - |
|
|
|
只有小组成员才能发言,加入小组>>
1278 浏览 0 评论
2371 浏览 1 评论
2187 浏览 1 评论
2061 浏览 5 评论
2945 浏览 3 评论
1096浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
747浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
916浏览 0评论
1279浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-21 00:34 , Processed in 1.321168 second(s), Total 51, Slave 44 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号