我正在尝试返回6611C上遥感线的电压。
使用+/-和RS +/-的后连接器4针连接进入负载,无论远程/本地开关设置如何,该命令始终返回相同的值。 设置:设定点为3V。 电阻器负载与两对导线连接。 切换到本地模式。 独立的DMM测量RS +/-引脚。 运行测试时,CV模式下的电源。 远程检测电压2.8V使用“MEAS:VOLT:DC?” 返回3.0V(本地/远程开关没有区别)在远程模式下,验证电压变为3.0V。 我希望能够通过6611C测量远程感应电压,但它似乎没有返回预期值。 6611的程序员手册说这个查询返回电源检测端子的电压。 我没有看到它...命令用法有什么问题? Scott编辑:szastoupil于2014年11月11日下午4:02编辑:szastoupil于2014年11月11日下午4:03 以上来自于谷歌翻译 以下为原文 I'm trying return the voltage of the the remote sense lines on the 6611C. Using the rear connector 4 pin connections of +/-, and RS+/- going to a load, the command always returns the same value regardless of the Remote/Local switch setting. Setup: Setpoint at 3V. resistor load with both pairs of wires connected to it. Switch in Local mode. Independent DMM measuring the RS+/- pins. Power supply in CV mode when running test. Remote sense voltage 2.8V Using "MEAS:VOLT:DC?" returns 3.0V (local/remote switch makes no difference) Verified voltage goes to 3.0V when in Remote mode. I want to be able to measure the remote sense voltage via the 6611C, but it doesn't seem to be returning the expected value. The programmer's manual for the 6611 says this query returns the voltage at the power supplies sense terminals. I'm not seeing it...what is the issue with the command usage? Scott Edited by: szastoupil on Nov 11, 2014 4:02 PM Edited by: szastoupil on Nov 11, 2014 4:03 PM |
|
相关推荐
1个回答
|
|
您好斯科特,您的设置和测量情况对我来说有点不清楚。
我对2.8V和3.0V感到困惑。 但是,您似乎已将输出电压编程(设置?)至3.0V。 另外,我认为你的意思是你已经证实在外部远程放置的电阻上测得的电压也是3.0V(?)。 如果我已正确理解这一点,6611C的功能与设计相同。 在远程感应模式下(在6611C的规格范围内),仪器应将实际输出电压(跨越远程负载)保持在编程值,其中±Sense线连接到±输出引线。 这也是MEAS:VOLT:指令适用的位置。 MEAS:VOLT不返回可能出现在6611C后面板上的±Sense连接器/插头上的电压值。 此致,PeterW 以上来自于谷歌翻译 以下为原文 Hello Scott, Your set-up and measurement situation is a bit unclear to me. I'm a little confused with the 2.8V and 3.0V. However, it appears that you have programmed (set?) the output voltage to 3.0V. Additiionally, I intrepret your meaning to be that you have verified that the voltage measured across your external, remotely placed resistor is also 3.0V (?). If I have understood this correctly, the 6611C is functioning as designed. In the Remote Sense mode (within the specification of the 6611C) the instrument should maintain the actual output voltage (across the remote load) at the programmed value, where the ±Sense lines attach to the ±Output leads. This is also the location where the MEAS:VOLT: directive applies. MEAS:VOLT does not return the value of the voltage which might appear at the ±Sense connector/plug on the rear panel of the 6611C. Regards, PeterW |
|
|
|
只有小组成员才能发言,加入小组>>
1226 浏览 0 评论
2348 浏览 1 评论
2159 浏览 1 评论
2024 浏览 5 评论
2905 浏览 3 评论
970浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
703浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
804浏览 0评论
1226浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-11-24 16:48 , Processed in 1.276373 second(s), Total 78, Slave 62 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号