是否可以将触发输出极性更改为下降沿而不是8164B上的上升沿?
我正在扫描一个81600B可调谐激光器,我想在扫描期间输出下降沿触发而不是上升沿。 8164B手册的第69页似乎表明默认模式是上升沿。 “外部触发连接器的触发电平默认为高电平......”我希望这意味着我可以将其更改为下降沿,但我没有找到配置极性的命令或设置。 谢谢你,Brian 以上来自于谷歌翻译 以下为原文 Is it possible to change the trigger output polarity to falling edge instead of rising edge on the 8164B? I am sweeping an 81600B tunable laser and I would like to output falling edge triggers during the sweep instead of rising edge. Page #69 of the 8164B manual seems to indicate that the default mode is rising edge. "The trigger level at the external trigger connectors is by default active high..." I am hoping that means that I can change it to falling edge but I haven't found a command or setting to configure the polarity. Thank you, Brian |
|
相关推荐
1个回答
|
|
|
嗨布莱恩,我不认为有改变这个输出触发极性的功能。
如果有必要,可能需要外部电路。 麦克风 以上来自于谷歌翻译 以下为原文 Hi Brian, I don't think there is function to change this output trigger polarity. Probably if that is necessary, then an external circuit will be needed. Mike |
|
|
|
|