您好,我正在使用固件版本为A.90.09.09固件的67GHz PNA-X。
我正在尝试使用梳状滤波器发生器来执行55MHz至33GHz的相位参考校准。 我可以毫无困难地对源衰减值0,10和20dB进行校准。 然而,当我达到30dB时,我分别在校准的步骤5和7中得到了“不均匀的源”错误。 (见附件)。 这也发生在40和50dB的源衰减。 当我尝试将此特定的第1层校准用于相位与驱动测量时,我的增益测量值比预期值高出约3dB。 如果有人可以让我知道为什么我会收到此错误以及是否可以修复,那将非常感激。 谢谢,rok comb filter cal error src atten = 30db step 7.tiF260.9 KB 以上来自于谷歌翻译 以下为原文 Hello, I am using a 67GHz PNA-X with firmware version A.90.09.09 firmware. I am trying to use the comb filter generator to perform a phase reference calibration from 55MHz to 33GHz. I had no trouble performing the calibration for source attenuation values of 0,10, and 20dB. However, when I got to 30dB, I got 'unleveled source' errors in step 5 and 7 of the calibration respectively. (see attachments). This also happened for source attenuation at 40 and 50dB as well. When I try to use this particular Tier 1 calibration for a phase vs drive measurement, my gain measurement is higher by about 3dB from expected. If someone can let me know why I am getting this error and whether it is fixable or not, that would be greatly appreciated. Thanks, rok 附件 |
4个回答
您可能想尝试更新的固件。
从A.09.90.11发布说明“固定:相位参考校准失败(源不平衡和接收器过载),30dB源衰减。” 这里的最新客户版本 - > http://na.tm.agilent.com/pna/firmware/firmware.html是A.09.90.12问候 以上来自于谷歌翻译 以下为原文 You may want to try newer firmware. From A.09.90.11 release notes "Fixed: Phase-reference calibration fails (source unlevel and receiver overload) with 30dB source attenuation." The latest customer release here --> http://na.tm.agilent.com/pna/firmware/firmware.html is A.09.90.12 Regards |
linlin10 发表于 2018-11-22 10:57 你好,谢谢你的回复! 我会尝试,看看是否能解决问题。 韩国 以上来自于谷歌翻译 以下为原文 Hi, Thanks for the reply! I'll try that and see if that fixes things up. rok |
没问题,希望对你有用。 就个人而言,我一直只使用一个具有0dB源衰减的相位参考校准。 然后,在校准需要相位校正的通道时,我将在该通道上执行全部校准,这会在校准期间提示单独的功率校正,并允许将单相参考校准应用于具有不同衰减器设置的通道。 以上来自于谷歌翻译 以下为原文 No Problem, hope that works for you. Personally, I have been using just one phase reference cal with 0dB source attenuation. Then, when calibrating a channel that requires phase correction, I will perform a cal all on that channel, which prompts for separate power correction during calibration and allows for the one phase reference cal to be applied to channels with different attenuator settings. |
linlin10 发表于 2018-11-22 11:26 嗨再次,我刚刚安装了A.09.90.12固件,并确认在相位参考梳状滤波器校准期间,在源衰减水平为30dB或更高时,它确实解决了未发现的源错误。 不幸的是,它没有修复噪声系统GUI错误,我将在一个单独的消息中发布。 再次感谢,rok 以上来自于谷歌翻译 以下为原文 Hi again, I just installed the A.09.90.12 firmware and verified that indeed it fixed the bug with the unleveled source error during the phase reference comb filter calibration at source attenuation levels of 30dB or higher. Unfortunately, it didn't fix a noise figure GUI bug, which I'll post in a separate message. Thanks again, rok |