你好,我的E3640A我观察到以下内容。
关闭输出后,我仍然在输出端测量23 mV的电压。 启用输出时,此偏移量仍然存在并根据此表更改:set [V] / measured [V] -------------------------- --------- off / 0.028 0 / 0.028 1 / 1.026 2 / 2.024 3 / 3.022 4 / 4.022 5 / 5.020 6 / 6.019 7 / 7.017 8 / 8.016有没有办法调整/纠正这种行为? 谢谢你的建议! 最好的问候彼得 以上来自于谷歌翻译 以下为原文 Hello, with my E3640A I observe the following. With the output switched off I still measure a voltage of 23 mV at the output. When enabling the output this offset remains present and changes according to this table: set [V] / measured [V] ----------------------------------- off / 0.028 0 / 0.028 1 / 1.026 2 / 2.024 3 / 3.022 4 / 4.022 5 / 5.020 6 / 6.019 7 / 7.017 8 / 8.016 Is there any way to adjust / correct this behavior ? Thank you for any advice! Best regards Peter |
|
相关推荐
1个回答
|
|
您好,我通过本手册中描述的“一般校准/调整程序”解决了这个问题。
关心彼得 以上来自于谷歌翻译 以下为原文 Hello, I solved the issue by going through the "General Calibration/Adjustment Procedure" described in the manual. Regards Peter |
|
|
|