2013年3月13日*信号分析仪基础知识和新应用网络广播*的问题和答案*问题:*有人提到参考电平设置因素在早期的幅度精度,但数字IF处理(A / D)不是一个因素?
以上来自于谷歌翻译 以下为原文 Questions and Answers to *Signal Analyzer Fundamentals and New Applications Webcast* on 13 March 2013 *Question:* It was mentioned that reference level setting factors into amplitude accuracy early on, but with digital IF processing (A/D) it's not a factor? |
|
相关推荐
1个回答
|
|
*答案:*较旧的模拟频谱分析仪具有一个对数放大器,该对数放大器在参考电平校准,并且由于其非线性响应而在参考电平以下测量的电平具有误差。
现代信号分析仪(PSA,PXA,MXA,EXA,CXA,MXE)具有极其线性的数字IF。 线性度从大约dB提高到小于百分之一dB。 这种改进类似于您从模拟盒式磁带转换到光盘的过程。 测量信号的数字化比模拟测量更准确和可预测。 重要的是要记住,改变参考电平会导致衰减的变化,这也会影响幅度精度。 以上来自于谷歌翻译 以下为原文 *Answer:* Older analog spectrum analyzers had a log amplifier that was calibrated at the reference level and had error for levels measured below the reference level due to its non-linear response. Modern signal analyzers (PSA, PXA, MXA, EXA, CXA, MXE) have extremely linear digital IFs. The linearity has improved from about a dB to less than a hundredth of a dB. The improvement is similar to what you would see transitioning from analog cassette tapes to compact discs. A digitization of the measured signal is much more accurate and predictable than an analog measurement. It is important to remember that changing the reference level can result in a change of attenuation which would affect amplitude accuracy, too. |
|
|
|
只有小组成员才能发言,加入小组>>
1296 浏览 0 评论
2377 浏览 1 评论
2199 浏览 1 评论
2068 浏览 5 评论
2955 浏览 3 评论
1131浏览 1评论
关于Keysight x1149 Boundary Scan Analyzer
759浏览 0评论
N5230C用“CALC:MARK:BWID?”获取Bwid,Cent,Q,Loss失败,请问大佬们怎么解决呀
939浏览 0评论
1296浏览 0评论
小黑屋| 手机版| Archiver| 电子发烧友 ( 湘ICP备2023018690号 )
GMT+8, 2024-12-27 05:18 , Processed in 1.341184 second(s), Total 77, Slave 61 queries .
Powered by 电子发烧友网
© 2015 bbs.elecfans.com
关注我们的微信
下载发烧友APP
电子发烧友观察
版权所有 © 湖南华秋数字科技有限公司
电子发烧友 (电路图) 湘公网安备 43011202000918 号 电信与信息服务业务经营许可证:合字B2-20210191 工商网监 湘ICP备2023018690号