  I've been having an issue with the Partial Calibration on the 4-port ENA series analyzer.

If I complete the following scenarios:

Scenario 1
----------------

- Perform a 4-port cal but make a mistake on the reflection part of the cal, the Thru is connected correctly though.

- Perform a partial cal, re-doing the reflection (Open/Short/Load) correctly


Scenario 2
---------------

- Perform a 4-port cal but make a mistake on the thru part of the cal, the Reflection is connected correctly though.

- Perform a partial cal, re-doing the Thru cals correctly


In scenario 1, the final calibration is always bad, but in scenario 2 it is good Why is this?

It seems like a good Reflection calibration always has to have been performed, before a Thru cal can be correct.  

回帖(1)

孔德羲

2019-9-19 16:46:40
我不太熟悉ENA的内部细节,但这让我感到惊讶。
使用PNA,可以获取并存储所有原始数据,以便校准可以按任何顺序进行,最终结果根据最后取得的值计算得出。
也许ENA首先计算1端口错误条款并且不保存原始测量值,这可以解释为什么单端口测量中的错误传播到最终校准。编辑:Dr_joel于2013年1月21日下午9:28

以上来自于谷歌翻译


     以下为原文

  I'm not so familiar with the ENA internal details, but this surprises me.  With the PNA, all the raw data is taken and stored, so that the calibration can proceed in any order and the final result computed from the last values taken.  Perhaps the ENA first computes the 1 port error terms and doesn't save the raw measurements, this would explain why errors in one-port measurements propogate to the final cal.

Edited by: Dr_joel on Jan 21, 2013 9:28 PM