有故障的34410A

我的34410A万用表有问题。
在某些范围内,它适当地测量,在一些范围内具有一点偏移。
我试着校准它。
它通过了ADC校准,但未通过DMM校准。
我得到错误707校准错误:在第一个DMM校准步骤中信号测量超出范围,这是零调整。
我进行了自检并得到错误623 ADC Mesure Zero Overload Test。
服务手册中未提及该错误。
我试图修复它,但如果没有万用表操作的原理图和知识就很难。
我检查了所有可能的组件,它们看起来很好。
是否有可能获得与导致该错误的电路相关的34410A原理图的至少一部分或某些故障排除提示?

以上来自于谷歌翻译


     以下为原文

  I have a problem with my 34410A multimeter. In some ranges it measures properly,
in some with a little offset. I tried to calibrate it. It passed the ADC calibration but it fails
the DMM calibration. I get error 707 Calibration error: signal measurement out of range
in the first DMM calibration step which is zero adjustment.

I ran a self-test and got error 623 ADC Mesure Zero Overload Test. That error is not
mentioned in the service manual.


I tried to repair it  but it's difficult without a schematic and knowledge of the multimeter
operation. I checked all the components I could and they seemed to be fine.
Is it possible to get at least a part of the 34410A schematic which is associated
with the circuitry causing that error or some troubleshooting tips?  

回帖(2)

郝汉

2019-2-18 16:39:27
不支持34410A的组件级修复。
原理图附有。
错误623 ADC测量零过载测试似乎表示自动归零操作失败。
对不起,我忍不住多了。
应将34410A送至安捷伦服务中心。

以上来自于谷歌翻译


     以下为原文

  Component level repair for the 34410A is not supported.  The schematics are attached.  Error 623 ADC Measure Zero Overload Test seems to indicate an autozero operation failure.  I'm sorry I can't help more.  This 34410A should be sent in to an Agilent Service Center.

附件

举报

周勇

2019-2-18 16:51:48
引用: lanziji 发表于 2019-2-18 20:42
不支持34410A的组件级修复。
原理图附有。
错误623 ADC测量零过载测试似乎表示自动归零操作失败。

非常感谢你。
我会试一试并试图修理仪表

以上来自于谷歌翻译


     以下为原文

  Thank you very much. I'll give it a try and attempt to fix the meter