好吧,你可能用你的英语欺骗了我:)我相信你所描述的问题是从恒定波切换到扫描模式时的不连续性。
Sweep必须进行大量的硬件配置,而且交换机从来就不是相位连续的。
但是,我建议您关闭输出,直到完成为应用程序配置框。
基本上等到发送SOURce1:FREQuency:MODE SWE后,打开输出。
这样你就可以避免不连续性。
如果扫描是总线或手动触发,那么它将像CW模式一样停在起始频率。
然后触发器使扫描开始并且是相位连续的,如前所述。
如果您还没有发现它,您可以使用扫描保持时间参数来停留在结束频率(如果您愿意),并且返回时间参数可以平滑地扫描回起始频率。
如果这不能满足您的需求,请告诉我。
使用测序还有另一种方法可以做到这一点,但它有点困难。
此致,卢克
以上来自于谷歌翻译
以下为原文
Well you could have fooled me with your English :)
The issue I believe you are describing is the discontinuity when switching from Constant Wave to sweep mode. Sweep has to do a lot of hardware config, and the switch was never meant to be phase-continuous.
However, I would suggest leaving your output off until you have finished configuring the box for your application.
Basically wait until after you have sent SOURce1:FREQuency:MODE SWE, to turn on the outputs. This way you will avoid the discontinuity. If the sweep is bus or manual triggered then it will park at the start frequency just like CW mode. The trigger then causes the sweep to start and is phase-continuous, as you saw previously.
If you hadn't discovered it already you can use the sweep hold time parameter to dwell at the end frequency if you'd like, and the return time parameter to smoothly sweep back to the start frequency.
If this doesn't satisfy your needs let me know. There is another way to do this using sequencing, but it is a little more difficult.
Regards,
Luke
好吧,你可能用你的英语欺骗了我:)我相信你所描述的问题是从恒定波切换到扫描模式时的不连续性。
Sweep必须进行大量的硬件配置,而且交换机从来就不是相位连续的。
但是,我建议您关闭输出,直到完成为应用程序配置框。
基本上等到发送SOURce1:FREQuency:MODE SWE后,打开输出。
这样你就可以避免不连续性。
如果扫描是总线或手动触发,那么它将像CW模式一样停在起始频率。
然后触发器使扫描开始并且是相位连续的,如前所述。
如果您还没有发现它,您可以使用扫描保持时间参数来停留在结束频率(如果您愿意),并且返回时间参数可以平滑地扫描回起始频率。
如果这不能满足您的需求,请告诉我。
使用测序还有另一种方法可以做到这一点,但它有点困难。
此致,卢克
以上来自于谷歌翻译
以下为原文
Well you could have fooled me with your English :)
The issue I believe you are describing is the discontinuity when switching from Constant Wave to sweep mode. Sweep has to do a lot of hardware config, and the switch was never meant to be phase-continuous.
However, I would suggest leaving your output off until you have finished configuring the box for your application.
Basically wait until after you have sent SOURce1:FREQuency:MODE SWE, to turn on the outputs. This way you will avoid the discontinuity. If the sweep is bus or manual triggered then it will park at the start frequency just like CW mode. The trigger then causes the sweep to start and is phase-continuous, as you saw previously.
If you hadn't discovered it already you can use the sweep hold time parameter to dwell at the end frequency if you'd like, and the return time parameter to smoothly sweep back to the start frequency.
If this doesn't satisfy your needs let me know. There is another way to do this using sequencing, but it is a little more difficult.
Regards,
Luke