嗨,我目前正在使用A.09.42.12固件。
我有一个.csa文件,我想将所有测量通道的组触发器设置为10。
但是,每当我回想起该文件时,它都会恢复为1.因此,由于某种原因,该设置未被保存。
这是一个痛苦,因为我有大约30多个频道,我必须一个接一个地将它改回10。
此外,以下是我遇到的一些其他小固件问题:1)在IM频谱测量类中,如果指定10kHz范围,则右边列出的频率不正确。
(详见附件)。
例如,它应该是999.995Mhz到1.000005 Ghz,而不是1.00001Ghz。
2)当使用具有较旧固件版本的PNA时,它将列出生成pdf打印输出时“跟踪属性表”中使用的平滑量。
但是使用新固件,这些信息似乎已经消失了!
3)对于具有多个标记的单个迹线,如果我对其启用跟踪统计,它将仅显示其中一个标记,并用统计表覆盖其余部分。
我不明白为什么你不能同时显示所有标记和统计表。
4)使用向导使用e-cal单元进行校准时,在某些步骤中,我注意到端口上的蓝灯已经点亮,甚至有机会点击“测量”按钮并将其连接到e-
cal端口!
我担心这可能会造成一些辐射危险,特别是在高功率和频率水平校准时。
如果有人可以查看触发设置未被保存的组并解决其他小问题,那将非常感激。
韩国
以上来自于谷歌翻译
以下为原文
Hi,
I'm currently using A.09.42.12 firmware. I have a .csa file where I want to to set the group trigger to 10 for all the measurement channels. However, every
time I recall the file, it reverts back to 1. So that setting is not being saved for some reason. It is a pain, since I have about 30+ channels and I have to change it back to 10 again one by one.
Also, here are some other minor firmware issues I have encountered:
1) In the IM spectrum measurement class, if you specify a 10kHz span, the frequency listed at the right edge is incorrect. (see attached for details). For the example, it should be 999.995Mhz to 1.000005 Ghz, not 1.00001Ghz.
2) When using the PNA with the older firmware version, it would list the smoothing amount used in the 'Trace Attributes Table' when generating a pdf printout. But with the new firmware, that information seems to have disappeared!
3) For a single trace with several markers, if I enable trace statistics on it, it will only show 1 of the markers and cover the rest up with the statistics table. I don't see why you can't show all the markers and the statistics table at the same time.
4) When performing a calibration with an e-cal unit using the wizard, during some steps, I noticed the blue light on the ports already lit on before even having a chance to click the 'Measure' button and connecting it to the e-cal port! I am worried this might pose some radiation risk, especially when calibrating at high power and frequency levels.
If someone can looking into the group triggering settings not being saved and address the other minor issues, that would be greatly appreciated.
rok