Hi,
I got my brand new 34465A DMM last week, it had been calibrated by Keysight (ini
tial/ production calibration) June 1st.
Out of the box, it had a disturbingly high calibration count of 222.
Another owner of a new 34470A found 76 cal counts, which seems also too high, maybe up to 20 or 40 seems to be reasonable, as the count seems to increase by 1 per calibration group only. (calibraton of 100V and 500V DCV increases by 1 only)
I assume, Keysight currently has big problems to finally calibrate these instruments in their production run.
I found out, that the DCV ranges 100mV, 1V, and +/- 10V were spot on, after using ACAL, always 10GOhm).
If I switch to 10MOhm, the reading is +3ppm higher in both ranges, although there is no hardware switching involved, as the HV ranges are already in this 10MOhm mode!
If I calibrate with 10MOhm setting, the verification also fails by +3ppm, switching to AUTO removes this error.
Both errors sum up to +17 and +20ppm, which already is the 90 day drift.
During Keysights calibration process, as ACAL probably is engaged before verification, that will probably cause a calibration error, explaining the high 222 counts of my instrument.
Secondly, this instrument will fail the 90 days or 1 year specification, if it already starts at 20ppm and the normal drifts will add to that.
So I need support from Keysight engineering, to get a confirmation of that bug, and a fix in the firmware, or in hardware, if it turns out to be HW related (less probable to my understanding).
I could also supply serial number, test report and further details of my own measurements.
Regards
Dr. Frank