是德科技
直播中

王秀华

7年用户 154经验值
私信 关注
[问答]

aperature/autozero如何影响记录的值

嗨,我希望有人能帮我理解aperature / autozero如何影响记录的值。
我想使用34461A来测量平均1s周期内的平均电流消耗。
被测设备(DUT)将非常灵活(例如,1uA通常会为几个我们提供10mA电流)。
我很高兴设置/钳制DMM上的当前范围(无自动量程问题)Q1。
我是否正确地认为Aperature将提供当前的真实平均值/积分?
Q2。
在自动归零期间我是否会丢失测量值?
例如
假设aperature设置为1s且自动调零。
100秒后,我将有100个读数。
这些读数的总和/ 100 = 100s期间的真实平均值?
或者,当DMM进行自动调零时,DUT的某些电流尖峰是否会丢失?

以上来自于谷歌翻译


     以下为原文

  Hi,

I'm hoping someone can help me understand how aperature/autozero affects the recorded values.
  
I would like to use a 34461A to measure average current consumption averaged over periods of 1s.  
The device under test (DUT) is going to be quite spikey (say 1uA typically bursting to a 10mA for few us).
I'm happy to set/clamp the current range on the DMM (no auto range issues)

Q1. Am I correct in thinking that the Aperature will provide a true average / integration of the current?

Q2. Do I loose measuements during autozero? e.g. say the aperature is set to 1s with auto zero on. After 100s, I will have 100 readings. Does the sum of these readings / 100 = true average for the 100s period? Or is it possible that some current spikes from the DUT will have been lost while the DMM was doing an auto zero?  

回帖(1)

王秀华

2018-9-25 14:59:26
刚刚通过电话与Keysight USA的应用工程师交谈。
在这里发布回复,以防其他人有帮助。
Q1:是德科技:“是的,这是正确的。每个结果都是电流窗口上电流的真正积分。”
我:很好,这就是我的想法。
问题2. Q2:是德科技:“在自动调零周期内没有读数丢失。”
我:我还是不确定。
但是真的不想在国际电话上闲逛和讨论。
无论如何,我对Q1更感兴趣。
要使用Autozero-Once确定一次。

以上来自于谷歌翻译


     以下为原文

  Just spoke to an apps engineer at Keysight USA by phone. 
Posted the response here in case it is helpful to someone else. 

Q1: 
Keysight: "Yes this is correct. Each result is a true integral of current over the aperature window.".
Me: Great, it's what I thought. On to Question 2. 

Q2: 
Keysight: "No readings are 'lost' during the auto zero cycle." 
Me: I'm still not sure about this. But didn't really want to hang around and discuss on an international call. I was more interested in Q1 anyway. Going to use Autozero-Once to be sure.
举报